February 21, 2017 04:36

I have a system that show a U01, U02,U03 error code when powered on. Any suggestions? Looking for some service information/ manual.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: MUX-200D-XC error codes

February 22, 2017 10:22

These are user errors referencing batteries you can reset them if you replaced or tested the batteries
U01 Battery Displayed after 2 years from installation.
U02 CPU Backup Battery Displayed after 2 years from installation.
U03 Emergency Brake Release Battery
Displayed after 2 years from installation.


Confirming the Usage Period, Count and Distance Information for Each
Part
Turn the key switch to ON while holding the remote control button down, to display the usage
period, count, and distance information for each part in the kV and mAs displays on the X-ray
control panel.
Toggle between information for each part by pressing the kV [+] and [-] buttons, or the region
button, to display the information. This blinks the lamp for the region button corresponding to the
selected replacement part.
Temporarily Disabling the Replacement Request Display Function
Use the following procedure to temporarily disable the replacement request display function.
1 Turn the key switch ON while pressing the remote controller button on the
X-ray control panel,Check the data on usage, number of cycles, and
distance for each part.
2 Press the mAs [+] and [-] buttons.
This displays [clr] [ no] in the kV and mAs displays on the X-ray control panel.
3 Press the mAs [+] and [-] buttons.
This displays [clr] [yes] in the kV and mAs displays on the X-ray control panel.
4 With [clr] [yes] shown on the kV and mAs displays, hold the registration
button down for at least 3 seconds.
This can temporarily disable replacement request display function.
To clear and reset go to installation manual and Appendix B

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)