MX8000 has problem. what is 'dose right' function

Tags: Test Equipment and Tools

kang seokhun

MX8000 has problem. what is 'dose right' function

February 13, 2014 01:15

dear sr.

we are in big trouble that use a MX8000 QUAD

this problem is 'tube current test failure'

we shut( expose ) protocol
120KV, 1.0Sec, 360degree, 100mAs(100mA)

but they measure tube current : 0.891VDC at 'IT act100' point

so we checked tube history in service parameter

current - nor : 870, act : 891

replaced E-box, HV tank but that measure not change

why shut low expose than set protocol ?
and do you have another solve?
please help sr.
wait for your mail tviori@naver.com