MX8000 has problem. what is 'dose right' function?

Tags: CT, PHILIPS

kang seokhun

MX8000 has problem. what is 'dose right' function?

February 18, 2014 09:04

Dear sr.

we are in big trouble that use a MX8000 QUAD

this problem is 'tube current test failure'

we shut( expose ) protocol
120KV, 1.0Sec, 360degree, 100mAs(100mA)

but they measure tube current : 0.891VDC at 'IT act100' point

so we checked tube history in service parameter

scan time-nor 1097, act 1030
current - nor : 870, act : 891

replaced E-box, HV tank but that measure not change

why shut low expose than set a protocol ?
and what is 'dose right' function?
please help sr.
wait for your mail tviori@naver.com

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: MX8000 has problem. what is 'dose right' function?

February 19, 2014 08:17

Have you scoped the Tube current Itact feedback and UI_Load

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: MX8000 has problem. what is 'dose right' function?

February 19, 2014 08:17

sorry the first message sent before I could finish the meassuring points are on the D400 in the CBox

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS Mx-8000 IDT/Brilliance 16 Slice CT Scanner -----Worldwide leader!!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)