Over 150 Missouri Auctions End Today - Bid Now

mx8000 i cant move table

Kalid Rasheed

mx8000 i cant move table

December 23, 2009 09:22

I have philips ct scan model mx8000 the table is
not move horezantal or vertical the light of gantry switched off when i press it at ferst error message appear ( timeout period expired not ready : master service ) please help me .

Log inor Register

to rate and post a comment

(28)

Veronica Alvarez

Re: mx8000 i cant move table

February 05, 2010 10:54

still have this problem????, can help respond email

Log inor Register

to rate and post a comment

(7)

tony ballista

re: Re: mx8000 i cant move table

October 03, 2012 06:15

Helo, one of mx8000 is working erratically the horizontal movement of the patient table. It always does fine the table top and fails intermittently the move of the table support. Any help on this pls?
Email: tballista@gmail.com.
Appreciated!!

Log inor Register