MX8000 PDC Problem

(32)
(2)

Jose Dominguez

MX8000 PDC Problem

July 08, 2013 11:19

After a Critical XRS 21 Error, I checked voltage at the slip ring where there suppose to be 560VDC and no voltage was found. I checked in the PDC and found that at K2 contactor I have 400VAC in the input but not at the other side (Contactor does not activate). Any help with this situation will be highly appreciated.

System is a Marconi MX8000 Quad CT.

Log inor Register

to rate and post a comment

(70)
(12)

Julio c. Gomez

re: MX8000 PDC Problem

July 09, 2013 06:42

check Stop report loop.

good luck.

Julio c. Gomez.
Veracruz, Mexico.

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: MX8000 PDC Problem

July 09, 2013 06:42

D450 or D451 maybe defective

Log inor Register

to rate and post a comment

(23)
(2)

Ed Perez

re: MX8000 PDC Problem

July 09, 2013 06:48

I have seen contractors go bad as also the Voltage sense to the contractor i think its 24volts

Log inor Register

to rate and post a comment