After a Critical XRS 21 Error, I checked voltage at the slip ring where there suppose to be 560VDC and no voltage was found. I checked in the PDC and found that at K2 contactor I have 400VAC in the input but not at the other side (Contactor does not activate). Any help with this situation will be highly appreciated.
GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449
Jose Dominguez
MX8000 PDC Problem
July 08, 2013 11:19
After a Critical XRS 21 Error, I checked voltage at the slip ring where there suppose to be 560VDC and no voltage was found. I checked in the PDC and found that at K2 contactor I have 400VAC in the input but not at the other side (Contactor does not activate). Any help with this situation will be highly appreciated.
System is a Marconi MX8000 Quad CT.
to rate and post a comment
Julio c. Gomez
re: MX8000 PDC Problem
July 09, 2013 06:42
check Stop report loop.
good luck.
Julio c. Gomez.
Veracruz, Mexico.
to rate and post a comment
Lin Dong
re: MX8000 PDC Problem
July 09, 2013 06:42
D450 or D451 maybe defective
to rate and post a comment
Ed Perez
re: MX8000 PDC Problem
July 09, 2013 06:48
I have seen contractors go bad as also the Voltage sense to the contractor i think its 24volts
to rate and post a comment