Mx8000 QUAD

February 20, 2015 04:28

Hi?
I'm facing a problem in the Mx8000 Quad CT system,
I can do tube conditionning, Air calibration but i can't make an exam,
after fill in the patient data and selcting protocol, i get ready hand in the screen to start exposer, but the CTBOX is not ready, the appropeiate button "hand" is not activated.

Someone has an experience with such problem?
Thank you.

2 people liked this
Log inor Register

to rate and post a comment

(34)
(2)

Haluk Aktas

re: Mx8000 QUAD

February 22, 2015 10:16

Email: hlkaktas@gmail.com

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

re: Mx8000 QUAD

February 22, 2015 10:17

The CT Box is broken or the Serial Conecction is missing. Check Serial Cables at the Operators Console. If there are OK, replace the CT Box.
Other option is a Software Problem. Try to reload the software.
Good Luck!

Log inor Register

to rate and post a comment

Mustapha EL ALLAOUI

re: re: Mx8000 QUAD

February 23, 2015 08:20

Thank you Gustavo for your help,
For your information, the CTBOX was replaced, but it donesn't solve the problem.

Log inor Register

to rate and post a comment

saju bhaskaran

re: Mx8000 QUAD

March 16, 2015 08:50

Contact on sajubhas@gmail.com

Log inor Register

to rate and post a comment

(19)

Badawy Naguib

re: Mx8000 QUAD

December 23, 2018 02:45

Dear najim i have the same problem
Can you help me

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)