Over 500 Total Lots Up For Auction at One Location - NJ 06/15

need help with philips bv libra

empeeno1 etlar

need help with philips bv libra

August 04, 2013 07:12

This bv libra c-arm cannot bootup. It displays 'Error in startup procedure 4001' Any help is appreciated with service manual also.

1 person liked this
Log inor Register

to rate and post a comment

Yuriy Kuznetsov

re: need help with philips bv libra

August 05, 2013 06:19

Communication error between DFI and MVS console (TRUIF). Restart the MVS. If this
problem remains, check the cabling between the DFI and the MVS console.

Log inor Register

to rate and post a comment

(21)
(2)

P402

re: need help with philips bv libra

August 06, 2013 09:51

4001 SCSI_E_READ
DFI disk failure.
CAUTION: After the following actions the PATIENT DATA,
DFI CONFIGURATION DATA and DFI IP SETS are LOST. Start the DFI power-up
service menu and low-level format the DFI disk. Re-install the DFI.
See service for more solutions

Log inor Register

to rate and post a comment

(47)
(7)
(6)

KONSTANTINOS PATTICHIS

re: re: need help with philips bv libra

December 09, 2015 12:59

Dear sir
I cannot load the software in the bv libra. I do not have the hw key. If you have a solution please tell me.

Thanks in advance
Pattichis Kostas

Log inor Register

to rate and post a comment

(42)
(7)

Saqib Mirza

re: need help with philips bv libra

September 10, 2013 08:52

Dear
if you still have the problem I can help you out.

Saqib

Saqib.mirzapk@gmail.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Certified Pre-Owned Endoscopes, All Major Brands, Best Pricing & Warranty.

We are the best choice for all your Endoscopy needs. We have Certified Pre-Owned refurbished and used Endoscopes, and all other related components and accessories. We are committed to your absolute satisfaction. (513) 469-7267

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)