Over 150 Missouri Auctions End Today - Bid Now

Needs Help Philips Optimus 50 Generator

Tags: Radiology - R/F, PHILIPS

(21)

(2)

can kutay orten

Needs Help Philips Optimus 50 Generator

November 27, 2012 09:15

Hi. I have a problem with Philips Bucky Diagnost with Optimus 50 generator. Everything seems OK but equipment is not shooting. There is no error code. Is there anyone can help me to solve the problem? I don't have any service manual or service soft. I think i need to connect to generator to upload generator configuration, seen error codes etc. Thanks in advance.

Log inor Register

to rate and post a comment

(8)

Gary Barnes

re: Needs Help Philips Optimus 50 Generator

November 27, 2012 07:44

If this is intermitent, software inhibits the exposure if more than four preps are completed within a minute or two to protect the anode, the user does not see the error however it is stored in the error log. It may be an interlock issue, have you got the green ready light on the CS, maybe try direct auxillary. Think basics like handswitch issue, if there was any corrupt config/cals the generator would not complete a succesful boot. Does the unit go into prep?

Log inor Register

to rate and post a comment

(21)

(2)

can kutay orten

re: re: Needs Help Philips Optimus 50 Generator

November 29, 2012 10:09

Hi Gary. When i press hand switch from generator, relay is getting the signal. So i don't think i have a problem with hand switch. But i am doubting generator is not getting ready. As i don't have any manual or experience for Philips x-ray, i am not sure what is the generator ready situation?

Log inor Register

to rate and post a comment

(8)

Gary Barnes

re: re: re: Needs Help Philips Optimus 50 Generator

November 30, 2012 09:33

The ready condition is when you have the green horse shoe on the generator console and green led on the CS.

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: Needs Help Philips Optimus 50 Generator

November 27, 2012 07:44

In order to have exposure enable the unit has to be in ready state indicated on the control console. Is it ready? If so, is there anode rotation when you press the exp. button? Why do you think that config is needed? Did the system work before normally? Is it a just installed system? Give more details.

Log inor Register

to rate and post a comment

(21)

(2)

can kutay orten

re: re: Needs Help Philips Optimus 50 Generator

November 29, 2012 10:09

System installed min.5 years before. I never took care of system. I was searching a service manual to understand better system. I asked some of my friends has Philips experience. Ex Philips engineer visited the system and he said this problem happened before and it needs generator config. one more time. For me generator is not getting ready, but to understand this i need service manual to learn how to see ready mode. Because, when i turn on system there is no error code. When i press hand switch relay from generator is getting signal so hand switch is OK. Generator control unit which is electronic board located top of generator are flashing red during start up after there is no any led visual or flashing. I don't have any idea ready condition.

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: re: re: Needs Help Philips Optimus 50 Generator

November 29, 2012 01:21

On the Operating Console there is a READY sign light(green) which you have to observe every time before you make an exposure.

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: Needs Help Philips Optimus 50 Generator

November 27, 2012 07:44

Are you using bucky or table top, if bucky, make sure the grid is moving

Log inor Register

to rate and post a comment

(21)

(2)

can kutay orten

re: re: Needs Help Philips Optimus 50 Generator

November 29, 2012 10:09

I tried all positions. But same. Basically, table top position i had to perform prep and exposure.

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: re: re: Needs Help Philips Optimus 50 Generator

November 29, 2012 01:21

Some more few things.... If you use any of the Bucky devices be sure you pull out and push back the selected Bucky tray completely before exposure attempt. This is a function to avoid double exposures on the same cassette.
Your system olso can have cassette sensing device so, without cassette you can not initiate exposure. On older systems this movement is also an essential mechanical charge of the catapult bucky system. This may happen if radiographer staff have changed and miss education of the followers.
Good luck

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: Needs Help Philips Optimus 50 Generator

November 29, 2012 01:21

It is very unlikely that the system loose coniguration without a sounding error message. Chose a direct radiation mode on the operating console ( which means no Iontomat Nor Bucky device are involved in the action)eg. wrist. Try to make exposure. Let me know if anode is rotating during the prep. phase and let me know what happens during exp. phase.
Still, you did not specify the system configuration.
Do not expect quick help without basic information of your trouble and I think, a service manual will not help you at all.

Log inor Register

to rate and post a comment

Ira Corbin

re: Needs Help Philips Optimus 50 Generator

December 13, 2012 10:39

Did you get this unit up? If it was working and stopped,
do the basics. Take out the buck tray and check for markers
behind it, preventing seating of the tray. Philips will not
shoot table top if the tray is pulled out. A door switch
should prevent "ready light" from coming on; bypass the
door switch contacts. Check the "second trigger" of the
exposure switch; and the cable if pendant switch is used.
Check the bucky interlocks to insure a "loaded bucky" is
not sensed by the generator when table top is selected.
Does the unit have cassette size sensing? Is it working?
Hope this helps.

Log inor Register

to rate and post a comment

(42)
(7)

Saqib Mirza

re: Needs Help Philips Optimus 50 Generator

January 05, 2013 10:59

Dear Can you a small half circle indication on the left of the console display, if its there it means generator in standy after booting and every thing is ok.
in this case pls select free cassette option for exposure.
Generator will not expose if there is no casste in Tray.
if no half circle there is some thing wrong in the generator, and error message also appears on the Console scree.

regards

1 person liked this
Log inor Register

to rate and post a comment

(1)

Muhammad Shahzad

re: Needs Help Philips Optimus 50 Generator

April 06, 2023 10:13

Same problem in my unit please help

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)