I have a problem with the PANI module, the problem is:
if the pump is conecting the pump don't turn on and don't have any voltage, but if I moved the pump from the conector, the voltage in the conector is 5V.
Someone can help me??
thanks
my email is palos182@hotmail.com
GCI offers complete MRI & CT sales & service with ROI analysis, in-house financing, contractor drawings, accreditation, training, marketing, and OEM-level support beyond the sale.
Carlos Palos
Nellcor NPB 4000 problem
September 30, 2009 04:52
HI
I have a problem with the PANI module, the problem is:
if the pump is conecting the pump don't turn on and don't have any voltage, but if I moved the pump from the conector, the voltage in the conector is 5V.
Someone can help me??
thanks
my email is palos182@hotmail.com
to rate and post a comment
Kadir AKMAN
Re: Nellcor NPB 4000 problem
November 04, 2009 01:19
do you hear noise from the transformator. change the capasitor (100uf 25v). system works normally.