Hi, I have an error messeage that " x ray source arcing failure". I need service manuals norland excell. my email is edward_f_m@hotmail.com or mendezpuntiel@gmail.com.
The X-Ray Source Arcing message means exactly what it says. This is usually caused by low oil in the tube and that loss of oil being replaced by air. You may find you can fill the tube by filling the bag and holding it above the tank to allow the air bubbles to go into the bag then pinch off the tube and refill the bag. This does take some time and many attempts to get it right. The tube will probably have to be replaced. The service manual will provide no information that can be of help here.
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Edward Mendez
Norland Excell Problem
June 06, 2015 04:54
Hi, I have an error messeage that " x ray source arcing failure". I need service manuals norland excell. my email is edward_f_m@hotmail.com or mendezpuntiel@gmail.com.
Thank
to rate and post a comment
Desmond Johnson
re: Norland Excell Problem
June 09, 2015 08:46
The X-Ray Source Arcing message means exactly what it says. This is usually caused by low oil in the tube and that loss of oil being replaced by air. You may find you can fill the tube by filling the bag and holding it above the tank to allow the air bubbles to go into the bag then pinch off the tube and refill the bag. This does take some time and many attempts to get it right. The tube will probably have to be replaced. The service manual will provide no information that can be of help here.
to rate and post a comment