Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

OEC 8800 interlock failure error code

(332)
(37)
(60)

Gary Benitez

OEC 8800 interlock failure error code

April 16, 2014 06:56

Does anyone know what the OEC 8800 interlock failure error code means.

Thanks,
Gary

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: OEC 8800 interlock failure error code

April 18, 2014 09:03

this can be several causes, fast stop switch curcuit open, the CPU not booting up (to check look for a green LED on the Monoblock control board if on CPU ok) 24 volt power supply bad or blown fuse on the Power Signal interface board F11 3.15 amp.
to test fast stops jumper P1 4 to 21 on the Power Signal interface board

1 person liked this
Log inor Register

to rate and post a comment

Willy La Torre

re: re: OEC 8800 interlock failure error code

May 02, 2014 04:44

Hi Lenny. I have a similar problem. I get the interlock failure message also but, F11 is good, 24V power supply is good, computer boots up. I dont get 15V on the power signal interface board, I have replace F8 several times, it keeps blowing. Do you know what the problem could be?
Thanks

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: OEC 8800 interlock failure error code

May 05, 2014 04:06

this feeds to several circuits turn unit off with your meter in ohms check between chassis ground disconnect connector J10 and ohm pin 1 to ground see if shorted, then check J11 pin 1 to ground, check J4 pin 6 to ground let me know if one is shorted

Log inor Register

to rate and post a comment

(2)

ibad sy

re: re: re: re: OEC 8800 interlock failure error code

September 05, 2014 09:39

hi lenny ...i am working on a OEC9800 it shows PRE CHARGE VOLTAGE ERROR ...in service manual it shows check F4 and F2 fuses or 24v for interlock but everything is fine the booting arrows stops at 15 after that the error comes .. please help me on this (briefly)waiting for your reply thanks in advance.....

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: OEC 8800 interlock failure error code

September 05, 2014 05:20

bad batteries to test place meter accross the black and red wires coming from battery pack and reboot if the DC voltage exceed 235VDC it is bad batteries

Log inor Register