OEC 9000 boot problem

Tags: C-Arm, OEC

(20)
(3)
(1)

Edwin Rosado

OEC 9000 boot problem

August 29, 2013 05:48

when i turn on the system the monitor cart display stay at B% and the C-ARM display at b7, any suggestion.

Log inor Register

to rate and post a comment

(22)
(6)

Mike Erlanson

re: OEC 9000 boot problem

August 30, 2013 06:53

It sounds like either your floppy drive or boot disk have gone bad. If I remember correctly if you take off the shell of the base of the C-Arm you should see a 3 1/2" floppy disk drive on the right side. You can test the drive to see if that is the issue. If it is the disk, then unless if you have a backup which I would think you most likely do not because those are ofter overlooked you might be up a creek.
If I am wrong on the location of the drive, it may be on the monitor cart, but I am prety sure it is onder the shell of the C-Arm.

Log inor Register

to rate and post a comment

(20)
(3)
(1)

Edwin Rosado

re: re: OEC 9000 boot problem

August 31, 2013 09:47

thanks, the monitor cart is B8, the monitor cart had a disk drive and the main frame had a floppy drive. i think my problem is in the monitor cart.