OEC 9000 Error B8

May 08, 2007 11:54

can some body help me? this machine an oec 9000 diasonics is stocked in the stage b8 I suppose can be the technique proc. board but I need be sure for change it ,exist any cause external that could generated this error ,please some body help is urgent... thanks joel

Log inor Register

to rate and post a comment

(110)
(10)
(3)

Deon Armes

Re: OEC 9000 Error B8

May 10, 2007 04:22

Joel,This comes up as a technique processor boot error code. However I have a list of things to check before replacing the TP board. 1.Check all supply voltages. 2.Check all connectors and cable and reseat them. 3.Check the time and date settings. 4.The book says you could have an eeprom mismatch(serial number mismatch). 5.Floppy drive/jumpers/switches. 6.Corrupted software diskette. Use the backup if you have one. 7.Replace TP board. Good luck!
Deon

Log inor Register

to rate and post a comment

(102)
(13)

Joel Grullon

Re: OEC 9000 Error B8

May 13, 2007 08:28

the eeprom mismatch ,the software into the diskette must match with the software into the eeprom ? , the question is because I allready changed the board and then I sure all is fine less the new board

Log inor Register

to rate and post a comment

(95)
(7)
(2)

Rick Ebling

re: Re: OEC 9000 Error B8

November 08, 2012 03:27

Deon , where are you located