hi Juan, I already checked the supply voltages and everything seems to be fine, but I keep gettting the B8 error in the display. And the monitor cart is already turned on and waiting for instruction, also I cant enter diagnostic menu, there is no response, hope you can give me some tips...
Thanks
The B8 in the Carm is not an error, it means is waiting for a response for the monitor cart, there are to lines coming in the main cable( the umbilical cord) tx and rx (transmit and received)pins 7 to 10 are you communication wires, check the main cable or the pins at the carm and going into the relay PCB ( the PCB inside the C behind the cable connectors)
Hi, I hope this is good. Mita I have a problem with a c-arm OEC 9000 the equipment starts without errors and generates x-rays and image in the monitor live but at the moment of releasing the switch or pedal of x ray or fluoro I generate the error E7 in the car of monitors , The communication is verified and is perfect, you can help me in indicating what the problem may be
We work with hospitals providing reliable ultrasound systems, probes & parts. By offering sales, service, rentals & repairs, we help our clients access high quality imaging systems while keeping costs low & support simple. Please call 760-834-4777!
Rick Ebling
oec 9000 lock up
November 08, 2012 03:21
OEC 9000 goes to B8 then locks up.
to rate and post a comment
Juan de la Cruz Martinez
re: oec 9000 lock up
November 08, 2012 06:08
is waiting for the monitor cart to load.
to rate and post a comment
Carlos Carbajal
re: re: oec 9000 lock up
September 02, 2013 06:27
hi Juan, I already checked the supply voltages and everything seems to be fine, but I keep gettting the B8 error in the display. And the monitor cart is already turned on and waiting for instruction, also I cant enter diagnostic menu, there is no response, hope you can give me some tips...
Thanks
to rate and post a comment
Juan de la Cruz Martinez
re: re: re: oec 9000 lock up
September 03, 2013 08:28
The B8 in the Carm is not an error, it means is waiting for a response for the monitor cart, there are to lines coming in the main cable( the umbilical cord) tx and rx (transmit and received)pins 7 to 10 are you communication wires, check the main cable or the pins at the carm and going into the relay PCB ( the PCB inside the C behind the cable connectors)
to rate and post a comment
CARLOS PAEZ
re: re: re: re: oec 9000 lock up
February 13, 2017 10:20
Hi.
Hi, I hope this is good. Mita I have a problem with a c-arm OEC 9000 the equipment starts without errors and generates x-rays and image in the monitor live but at the moment of releasing the switch or pedal of x ray or fluoro I generate the error E7 in the car of monitors , The communication is verified and is perfect, you can help me in indicating what the problem may be
Best Regards
Carlos Páez
to rate and post a comment
Chris Baklik
re: oec 9000 lock up
November 12, 2012 11:54
I have the service manual e-mail me crisb1955@aol.com
to rate and post a comment