Over 150 Missouri Auctions End Today - Bid Now

OEC 9400 010E SYSTEM ERROR

(4)
(2)

Joy Broyles

OEC 9400 010E SYSTEM ERROR

February 10, 2012 09:11

Our OEC 9400 is showing a 010E System Error upon startup occasionally. Cannot find what this error means anyway. Any help would be appreciated. Thanks!

Log inor Register

to rate and post a comment

(19)
(1)
(2)

Eduardo Firmiano

Re: OEC 9400 010E SYSTEM ERROR

February 10, 2012 12:21

Try to replace the floppy driver.

Log inor Register

to rate and post a comment

(177)
(20)
(3)

Chris Baklik

Re: OEC 9400 010E SYSTEM ERROR

February 13, 2012 07:52

Try using the backup diskette if you have one to see if it help e-mail me if you need help

Log inor Register

to rate and post a comment

(6)

Glen Ridley

Re: OEC 9400 010E SYSTEM ERROR

February 13, 2012 04:26

Theis error is a communication error between the control panel and the tech processor. Try checking the cable form the back plane to ther control panel---they do go bad. it coulkd be a bad control panel processor causing the issue.