OEC 9400 c-arm with "KV ON IN ERROR" Code

Tags: C-Arm

(333)
(37)
(60)

Gary Benitez

OEC 9400 c-arm with "KV ON IN ERROR" Code

March 11, 2015 07:36

I have an OEC 9400 c-arm with "KV ON IN ERROR" when it tries to boot up. Does anyone recognize this error code?

Log inor Register

to rate and post a comment

(97)
(14)

Donald Bradbury

re: OEC 9400 c-arm with KV ON IN ERROR Code

March 12, 2015 11:29

An educated guess. You might check your hand, foot, and control panel exposure switches. You can unplug the foot and hand switches to eliminate them but you will have to ring out the front panel switch. Most x-ray systems don't like to power up with an exposure command initiated.

Log inor Register

to rate and post a comment

(14)

Lars Aagaard

re: OEC 9400 c-arm with KV ON IN ERROR Code

March 12, 2015 11:29

According to my service manual that error is a generator operation error code. It means KV was sensed without X-Ray requested. This indicates a fault with the high voltage generator.