FranciscoG

OEC 9400 Display code A7 Starting Interlocks and BPlus charge

April 27, 2016 06:47

Hello, this OEC 9400 got the batteries replaced and when I tried to turn it back on, the display on the C arm shows code A7: Starting Interlocks and BPlus charge (from manual). And it doesn't go further than A7.
Once the batteries were replaced the system was plugged into the wall to charge the batteries for more than 48hrs.

Has somebody faced this issue before? what did you do?

Thanks

1 person liked this
Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: OEC 9400 Display code A7 Starting Interlocks and BPlus charge

April 29, 2016 01:17

a cable is disconnected or loose check on the back of the card cage one of the cables there is probably the cause

2 people liked this
Log inor Register

to rate and post a comment

(43)
(21)

randy oswalt

re: OEC 9400 Display code A7 Starting Interlocks and BPlus charge

April 29, 2016 01:21

Just wondering. Did you do filament calibrations after battery change. May be unrelated. But having changed batteries may have disconnected something by accident in area of the cables pulling.

2 people liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: OEC 9400 Display code A7 Starting Interlocks and BPlus charge

May 02, 2016 12:54

Hi Randy, no I haven't done any calibration after changing the battery packs, because the system won't come up. I will check again all the cables.

Log inor Register

to rate and post a comment

(86)
(11)

Chris Salberg

re: OEC 9400 Display code A7 Starting Interlocks and BPlus charge

April 29, 2016 01:21

Did you calibrate and test the battery charger?

1 person liked this
Log inor Register

to rate and post a comment

Javier De leon

re: OEC 9400 Display code A7 Starting Interlocks and BPlus charge

May 08, 2018 09:41

Hello Francisco, I have the same problems from the star before I change the batteries pack the system did not came up from a7 did you resolve your problems, because I still working on the problems, can u help me

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: OEC 9400 Display code A7 Starting Interlocks and BPlus charge

May 09, 2018 09:46

If you send me an email address I can send you the a diagram to help troubleshoot. This can be a switch/wiring issue possibly a power supply, battery, fuse...

1 person liked this
Log inor Register

to rate and post a comment

Javier De leon

re: re: re: OEC 9400 Display code A7 Starting Interlocks and BPlus charge

May 09, 2018 03:13

I all ready send u an email tanks for the help regards

Log inor Register

to rate and post a comment

Javier De leon

re: re: re: OEC 9400 Display code A7 Starting Interlocks and BPlus charge

May 09, 2018 03:13

Javierfda30@gmail.com

Log inor Register

to rate and post a comment

(1)

Thieken Luke

re: re: re: re: OEC 9400 Display code A7 Starting Interlocks and BPlus charge

March 06, 2019 10:32

I have the same issue a7 change technique processor and disk. Reseeded all connectors with no luck. Any help available?