Over 150 Missouri Auctions End Today - Bid Now

OEC 9600 workstation hard disk problem

Tags: C-Arm, GE

(7)

(1)

Wiesław Dolata

OEC 9600 workstation hard disk problem

July 29, 2013 08:52

I have a problem with the workstation. I noticed one thing. When all boards has been removed (image processor, etc.) of a computer motherboard the system boots normally. After installing them back will not boot and we can not save or open a picture from the second disk partition. Do you know if that is the reason?

Log inor Register

to rate and post a comment

(12)
(3)

Jean Marc Cocqueel

re: OEC 9600 workstation hard disk problem

July 30, 2013 08:26

Good morning, we have in the start test an erreur 253: erreur Bios, do you know why? and how to fix it?
Thank you.

Log inor Register

to rate and post a comment

(12)
(3)

Jean Marc Cocqueel

re: OEC 9600 workstation hard disk problem

July 30, 2013 08:26

Good morning,
we have during start test error 253: boot error.
do you know what it is? and how to fix it?
thank you.

Log inor Register

to rate and post a comment

(7)

(1)

Wiesław Dolata

re: re: OEC 9600 workstation hard disk problem

July 30, 2013 09:09

Hi,
Sorry, but i don't know how to fix it.
Service manual says:
"*Error 253 occurs when the system does not boot fully and displays the "system error
#253" message. This message is associated with a loss of serial communication.
If an error message, such as "system error 253 at 0000:01F0" is displayed, the numbers
indicate the memory location where software detected the error.

Log inor Register