Over 150 Missouri Auctions End Today - Bid Now

OEC 9600

Tags: C-Arm

(23)
(2)

MedicalTao

OEC 9600

January 25, 2017 07:43

I have a OEC 9600. The both monitors get a scrambled image then shuts off. Has anyone experienced this issue before?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: OEC 9600

January 26, 2017 09:56

Have you checked the power supply? Reseat the connectors on the Aux interface and video switching boards then retest

Log inor Register

to rate and post a comment

(14)
(9)

stefano bandini

re: OEC 9600

January 26, 2017 09:58

If you intend during power on i think the PC don't start up.

1 person liked this
Log inor Register

to rate and post a comment

(9)

angel fitch

re: re: OEC 9600

October 14, 2019 10:17

I have the same problem, the monitor Is not working. I thinks the PC Is the problem, couse i can access yo the 3.5" disk for make a diagnostic

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: OEC 9600

October 15, 2019 10:33

what does the c-arm show? do you see an message on the monitor lower right corner?

Log inor Register

to rate and post a comment

Yaqub

re: OEC 9600

January 30, 2017 10:42

you have service manual

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: OEC 9600

January 30, 2017 03:22

Yes send an email address

Log inor Register

to rate and post a comment

(23)
(7)

Robert McKinnell

re:OEC 9600

October 28, 2019 10:32

Video card/Monitor or the Power Supply. Those are your options. If you swap left and right, see if the issues follow to each Monitor. Then you know its the monitor.

Log inor Register

to rate and post a comment

advertisement