oec 9800 -abs bad operation

(102)
(13)

Joel Grullon

oec 9800 -abs bad operation

September 24, 2012 12:20

hello friends, I need somebody to tell me how solve a problem with the abs system in my oec 9800, because when I activate the abs the kv and ma goes to max setting, how if I don't have video signal but at the monitor is too brighted, have video and I need to decrease the technique for can see anything at the monitor.

best regards Joel

1 person liked this
Log inor Register

to rate and post a comment

(37)
(6)

John LoPresti

re: oec 9800 -abs bad operation

September 24, 2012 05:31

check first to see if the iris on the camera or the collimator are closing down which will drive the dose to max, are you actually getting any kv, ma or radiation? You may also have a broken wire in the main harness for the ABs signal...many things can cause this

Log inor Register

to rate and post a comment

(102)
(13)

Joel Grullon

re: re: oec 9800 -abs bad operation

September 25, 2012 01:20

I have the video correctly because when I decrease the kv and ma I can see video at the monitor, what signal take the video controller for increase the kv and ma

Log inor Register

to rate and post a comment

(56)
(2)
(8)

Jairo Manzaneda

re: re: re: oec 9800 -abs bad operation

March 02, 2016 05:58

Hi joel, greetings, please let me know if you solved this technical situation and how?, regards

Log inor Register

to rate and post a comment

(20)
(10)

Ty Murphy

re: oec 9800 -abs bad operation

September 24, 2012 06:02

I second the broken wire opinion. I've had to repair those many times. Thankfully OEC seems to recognize the poor design and includes some spare wires for your harness repair pleasure!

Log inor Register

to rate and post a comment

(9)
(3)

Adenilson Rizzi

re: oec 9800 -abs bad operation

February 22, 2022 10:23

Hi friend, I have a problem similar to yours. Maximum kv in abs mode and manual mode there is image. Do you know what the defect was?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)