OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

Tags: C-Arm, GE

(12)

(1)

Guery Diaz

OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

September 06, 2017 12:38

Hello

I have a problem with an OEC 9800 plus. The power on sequence is interrupted in the 5th arrow, but monitor screens coming up without problem. The workstation boot up completely with or without the C arm connected, and in both cases the system interface board shows "AA" on its display. I checked pins of the interconnect harness from WS to C arm, all looks OK. All Power Supply voltages are OK. All fuses are good too. I checked the ARCNET signal coming from WS (with an oscilloscope) on the Fluoro functions board (TP6 & TP7) and looks like in the manual. The Fluoro Functions Board shows "00" on its display and all leds are on. Please, any suggestions would be appreciated.

Regards

Log inor Register

to rate and post a comment

(8)

Jeffrey Smith

re: OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

September 07, 2017 10:36

Check the 3 volts battery on the single board computer ( SBC )

Log inor Register

to rate and post a comment

(118)
(16)

VENGE

re: re: OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

September 08, 2017 10:18

Thanks Jeffrey!. The battery on the Gen Board had 0.2V. I replaced it, but I still have the same problem. Is probable than the data on the GEN board was erased due to this?

Log inor Register

to rate and post a comment

(8)

Jeffrey Smith

re: re: re: OEC 9800 doesn

September 08, 2017 10:48

You will need to set the correct date and time. You need an external keyboard to do that

Log inor Register

to rate and post a comment

(118)
(16)

VENGE

re: re: re: re: OEC 9800 doesn

September 08, 2017 05:58

NO :-(. Do you have the procedure to do that?

Log inor Register

to rate and post a comment

(6)

Chuck Orr

re: OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

September 07, 2017 10:36

5th arrow means: Generator object tasks initialize. If boot-up fails on this arrow, check communications between workstation and c-arm. Check ARCNet. Verify that workstation completely boots. Possible corrupted ARCNet nodes, loose boards, interconnect cable bad. Check to see if you have 5.1 Vdc on TP3 of the Control Panel Processor PCB under the doghouse....if not adjust R51 on PS1.

Log inor Register

to rate and post a comment

(118)
(16)

VENGE

re: re: OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

September 08, 2017 10:18

I have 4.9 in TP3 of control panel PCB, that counts as good voltage?

Log inor Register

to rate and post a comment

(6)

Chuck Orr

re: re: re: OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

September 08, 2017 10:48

No, needs to be 5.1.....bet this is your problem.

Log inor Register

to rate and post a comment

(118)
(16)

VENGE

re: re: re: re: OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

September 08, 2017 05:58

Thanks Chuck!. I can adjust this voltage from PS2?

Log inor Register

to rate and post a comment

(6)

Chuck Orr

re: re: re: re: re: OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

September 10, 2017 09:17

R51 on PS1 located on swing out panel of carm.

Log inor Register

to rate and post a comment

(9)
(1)

Greg Hackworth

re: OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

September 07, 2017 10:38

Take the interconnect cable cannon plug apart and ensure each wire is making good contact. On c-arm side, lightly push in on each interconnect pin; none of them should move. Unless something is wrong, the monitors will always work when disconnected from c-arm, which is a good TS technique to decide if c-arm or WS is at fault.

Log inor Register

to rate and post a comment

(118)
(16)

VENGE

re: re: OEC 9800 doesn't boot. Sequence interrupted in 5th arrow

September 08, 2017 10:18

Thanks Greg, I checked pins of the interconnect harness from WS to C arm, all looks OK.

Log inor Register

to rate and post a comment