Hello my friends, I'am having a problem with a OEC 9800 C-arm. Everytime an image is taken the image seems to be good for an instant, but then it blurs and the final image is useless, like if it was out of focus. If someone can comment on this problem, it will be greatly appreciated.
Does it get out of focus during Fluoro or during Digital Acquisition? If during Fluoro it gets blur commonly it could be related to an II (image intensifier) problem or the camera as well as the ABC - usually the ma and Kvp goes high even without an object.
If Fluoro is OK but problem is on digital acquisition then it could be related to Iris opening or Kvp - ma factoring.
I appreciate everyones comments, the problem is both on digital acquisistion and fluro. I will further investigate following your everyones recomendations, I will update you if I have any progress on this problem.
Lightly used system available direct from the end user customer. MONARCH QUEST
Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
Javier Aguilera
OEC 9800 image blurring problem.
June 05, 2014 01:35
Hello my friends, I'am having a problem with a OEC 9800 C-arm. Everytime an image is taken the image seems to be good for an instant, but then it blurs and the final image is useless, like if it was out of focus. If someone can comment on this problem, it will be greatly appreciated.
Thank you all.
to rate and post a comment
Kamran Butt
re: OEC 9800 image blurring problem.
June 06, 2014 08:42
Dear Javier,
Try after disabling Automatic Brightness Control.
to rate and post a comment
gerardo millena
re: OEC 9800 image blurring problem.
June 06, 2014 08:43
Does it get out of focus during Fluoro or during Digital Acquisition? If during Fluoro it gets blur commonly it could be related to an II (image intensifier) problem or the camera as well as the ABC - usually the ma and Kvp goes high even without an object.
If Fluoro is OK but problem is on digital acquisition then it could be related to Iris opening or Kvp - ma factoring.
to rate and post a comment
stefano bandini
re: OEC 9800 image blurring problem.
June 06, 2014 08:43
I think the problem was a power supply of image intesifier. You ca try to make a regulation with a focus trimmer of p.s. put under the cover of IBV.
to rate and post a comment
Alex Kalish
re: OEC 9800 image blurring problem.
June 06, 2014 08:44
Does this happen in all mag modes? Are you switching between radiographic and fluoro? I would adjust the intensifier power supply. If not, replace it.
to rate and post a comment
Javier Aguilera
re: OEC 9800 image blurring problem.
June 06, 2014 12:20
I appreciate everyones comments, the problem is both on digital acquisistion and fluro. I will further investigate following your everyones recomendations, I will update you if I have any progress on this problem.
Thank you all!