Over 500 Total Lots Up For Auction at One Location - NJ 06/15

OEC 9800 iris pot replacement procedure

Tags: C-Arm, OEC

(50)
(4)

Douglas Gray

OEC 9800 iris pot replacement procedure

September 20, 2012 06:15

Does anyone have a procedure for replacing an Iris pot on the collimator for an OEC 9800?

Log inor Register

to rate and post a comment

(177)
(20)
(3)

Chris Baklik

re: OEC 9800 iris pot replacement procedure

September 21, 2012 08:22

I have the service manual e-mail me crisb1955@aol.com

Log inor Register

to rate and post a comment

(37)
(6)

John LoPresti

re: OEC 9800 iris pot replacement procedure

September 21, 2012 06:11

Your going to need the RUT software and cable to connect to the workstation to do the iris calibration here is a sample from the service manual adjustments which are done with the software-
1. Uncheck the Show Circular Mask box.
2. Click the Calibrate Mag Modes button.
3. At the Collimator Iris Calibration - MAG2 window, make a MAG2 Fluoro exposure and position the flats of the Collimator Iris blades between the C rings of the BAFT image that appears on the left monitor (see the following illustration).
4. At the Collimator Iris Calibration - MAG1 window, make a MAG1 Fluoro exposure and position the flats of the Collimator Iris blades between the B rings of the BAFT as it appears on the left monitor. (Refer to the previous illustration.)
5. Click Calibrate Normal mode.
6. At the Collimator Iris Calibration - NORMAL window, make a Normal Fluoro exposure and position the flats of the Collimator Iris blades until no more than 1.5 mm of any Iris blade is visible (only some blades may be visible). Click OK
when finished.
7. At the Collimator Leaves Calibration window, open collimator leaves until they are just outside the visible image. Click
OK.
The service manual may come in handy too...good luck!

Log inor Register

to rate and post a comment

(3)

Angelica Potts

re: re: OEC 9800 iris pot replacement procedure

November 01, 2017 09:19

How can I get my hands on the RUT software?

Thanks!
angelicap@ancorapain.com

Log inor Register

to rate and post a comment