Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

OEC 9800 ISSUES

Tags: C-Arm, GE OEC

(78)
(7)

R GOPAL

OEC 9800 ISSUES

April 03, 2017 01:24

Hi friends,

I am getting folowwing error messages when I run Diagnostics on boards.

1. System Interface Board.
Error -150 -MACRO FAIL - line 37 of C:\9800\DIAG SI\WS-PWR.MO1 Value contained in peak buffer if out of range.

2. Video Controller board.
Error -104 -MACRO FAIL -line 16 of C:\9800\DIAG\VC\TESTGAIN.MO1
ERROR COMPARING WORD VALUES.

3. Image Processor Board.
Error - 104 - MACRO FAIL - line 17 of
C:\9800\DIAG\IP\T2CMP1.MO1
Error comparing word values.

What is the significance of these error messages?
Machine functioning normally except for frequent booting problems.
Can anyone guide me.
Thank in advance.
R.Gopal
aparnaelectromech@gail.com

1 person liked this
Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: OEC 9800 ISSUES

April 04, 2017 11:10

Check the capacitors around the CPU IC on the SBC.

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: OEC 9800 ISSUES

April 04, 2017 12:29

The run time errors are divided into four groups:
1. I/O errors - 100-149
2. Critical errors - 150-199

Error 150 is a critical error and 140 is an I/O error

I would first check the hard disk, replace it if faulty

If disk is ok, I would then perform a software reload because it seems the system cannot locate some files or write some files to the disk

Log inor Register

to rate and post a comment

(4)

Zac Mountz

re: OEC 9800 ISSUES

April 04, 2017 12:33

Check the 5vdc power supply, adjust if needed. This causes a lot of issues.