OEC 9800 problem with X-Ray Generation

(17)
(2)
(1)

Dmitry Romashkovtsev

OEC 9800 problem with X-Ray Generation

August 10, 2013 08:16

Good day!
I have the biggest problem ever!
First of all i had dead IGBTs and fuse on snubber PCB, I've changed them. Then i've tryed to adjust dead time of IGBTs, but after that adjustment i had Low kV message and no image on the monitor. Then i' ve change the tube... Didn't help. Some time after i've got an image, i don't know how. BUT!!! Still Low kV message on C-Arm display, and i noticied that i have an image with all kV but not greater then 1,5 mA... Could some one help me somehow, i have no ideas left... My email: dicamomil@gmail.com

Log inor Register

to rate and post a comment

Charlie Britt

re: OEC 9800 problem with X-Ray Generation

August 12, 2013 10:59

Did you have any problem adjusting the X-ray tube? How about batteries? Have you checked the HV connector ( tube or hv tank side) check the pins.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: OEC 9800 problem with X-Ray Generation

August 12, 2013 11:01

have you tried to recalibrate the generator, i had a unit once that when the IGBT's failed it corrupted my calibration files after erasing the FFB and GEN nodes I restored my back-ups and it corrected the problem

Log inor Register