OEC 9800 problem

(34)
(4)
(3)

Thameen Ansari

OEC 9800 problem

January 24, 2012 09:10

Hi
I have a problem with OEC 9800, 3 months before, We had mA on error and low mA error. Recalibrated null adjustment and Filament calibration. But we experienced error when doing Filament calibarion and showed ' filament current can not reach the requirwd level' it happened couple of times. We replaced battery set and HV regulator board. Calibrated the Null adjustment on new HV regulator board. now intermittently mA on error coming.. please direct me what is the caiuse of problem..

Many Thanks

Thameen

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: OEC 9800 problem

January 24, 2012 10:53

I'd be very suspicious of the cable harness from the C-arm base to the tube/collimator and II/CCTV. I think there are a couple of extra wires in the harness that you may be able to use if you find a flaky wire. Last thing would possibly be the x-ray tube itself.

Log inor Register

to rate and post a comment

Quint Bridenstine

Re: OEC 9800 problem

January 24, 2012 05:34

I am fairly new to the oec 9800 but had to through my two cents in. I would look at the "gib" board .let me know when you get it

Log inor Register

to rate and post a comment

(37)
(6)

John LoPresti

Re: OEC 9800 problem

January 27, 2012 05:43

Check the HV cable ends for arcing or flakey wires, regrease and tighen. The other fellow also had a good idea about a partially broken wire in the cable harness running from the base up through the C which is also HV cable is to the tube.
Once you've exhaused all/everyones suggestions start suspecting the tube itself. You've replaced/calibrated a lot of logical things already

Log inor Register

to rate and post a comment

(1)

Eric Bennett

Re: OEC 9800 problem

January 24, 2012 05:34

I had a similar MA error that was very intermittent. I went through just about the same process you did and finally I concluded that the resister divider in the high voltage tank was the cause. It took some courage to order such an expensive part but when I replaced the HV tank the problem never returned. I assume you tried reseating the cannon plug that connects to the top of the HV tank.

Log inor Register

to rate and post a comment

CRAIG WAMBURA

Re: OEC 9800 problem

January 25, 2012 07:21

replace your filament boards. mA errors are mainly caused by filament board malfunction.

Log inor Register

to rate and post a comment

(34)
(4)
(3)

Thameen Ansari

Re: OEC 9800 problem

January 30, 2012 09:08

Hi Thank you very much for all your technical supports and it is very useful to direct me to solve the problem. I will come back to this forum with more information soon.. thanks again

Thameen

Log inor Register

to rate and post a comment

(6)

Glen Ridley

Re: OEC 9800 problem

January 27, 2012 05:50

There is a adjustment called the Ramp Peak Voltage adjustment--it must be 3.8 vdc---do this with a scope on TP57 on the HVSR.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We provide sales, service & repair for portable ultrasound systems & probes

If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)