I have an intermittent problem with an OEC9900. About 1 out of 10 times I get a "System Error, The workstation is fully functional but X-rays are disabled. To enable X-rays turn the system off, wait 10 seconds then restart the system" I have tried the FFB and GIB boards. The debug monitor shows "13B Error>FFB monitor echo message failed, then 2 lines down I get "GIB error>event 70 FFB recovery timeout" I get that every time it does fail. Between the times it works and the times it does not. I do not touch anything except the on/off button on the workstation.
We are the best choice for all your Endoscopy needs. We have Certified Pre-Owned refurbished and used Endoscopes, and all other related components and
accessories. We are committed to your absolute satisfaction. (513) 469-7267
Bug
OEC 9900
February 10, 2014 03:25
I have an intermittent problem with an OEC9900. About 1 out of 10 times I get a "System Error, The workstation is fully functional but X-rays are disabled. To enable X-rays turn the system off, wait 10 seconds then restart the system" I have tried the FFB and GIB boards. The debug monitor shows "13B Error>FFB monitor echo message failed, then 2 lines down I get "GIB error>event 70 FFB recovery timeout" I get that every time it does fail. Between the times it works and the times it does not. I do not touch anything except the on/off button on the workstation.
to rate and post a comment
Preston Hill
re: OEC 9900
February 10, 2014 05:47
Check power supplies in monitor cart and c-arm. 5 dc = 5.2vdc
to rate and post a comment
TIBS
re: OEC 9900
February 10, 2014 08:48
Reload software and check your 5 volt power supply.