Bug

OEC 9900

February 10, 2014 03:25

I have an intermittent problem with an OEC9900. About 1 out of 10 times I get a "System Error, The workstation is fully functional but X-rays are disabled. To enable X-rays turn the system off, wait 10 seconds then restart the system" I have tried the FFB and GIB boards. The debug monitor shows "13B Error>FFB monitor echo message failed, then 2 lines down I get "GIB error>event 70 FFB recovery timeout" I get that every time it does fail. Between the times it works and the times it does not. I do not touch anything except the on/off button on the workstation.

Log inor Register

to rate and post a comment

Preston Hill

re: OEC 9900

February 10, 2014 05:47

Check power supplies in monitor cart and c-arm. 5 dc = 5.2vdc

1 person liked this
Log inor Register

to rate and post a comment

(16)
(7)

TIBS

re: OEC 9900

February 10, 2014 08:48

Reload software and check your 5 volt power supply.
If you have further questions contact 1-866-499-3966
Ext. 101. or visit www.tibscorp.com...

1 person liked this