I have an intermittent problem with an OEC9900. About 1 out of 10 times I get a "System Error, The workstation is fully functional but X-rays are disabled. To enable X-rays turn the system off, wait 10 seconds then restart the system" I have tried the FFB and GIB boards. The debug monitor shows "13B Error>FFB monitor echo message failed, then 2 lines down I get "GIB error>event 70 FFB recovery timeout" I get that every time it does fail. Between the times it works and the times it does not. I do not touch anything except the on/off button on the workstation.
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Bug
OEC 9900
February 10, 2014 03:25
I have an intermittent problem with an OEC9900. About 1 out of 10 times I get a "System Error, The workstation is fully functional but X-rays are disabled. To enable X-rays turn the system off, wait 10 seconds then restart the system" I have tried the FFB and GIB boards. The debug monitor shows "13B Error>FFB monitor echo message failed, then 2 lines down I get "GIB error>event 70 FFB recovery timeout" I get that every time it does fail. Between the times it works and the times it does not. I do not touch anything except the on/off button on the workstation.
to rate and post a comment
Preston Hill
re: OEC 9900
February 10, 2014 05:47
Check power supplies in monitor cart and c-arm. 5 dc = 5.2vdc
to rate and post a comment
TIBS
re: OEC 9900
February 10, 2014 08:48
Reload software and check your 5 volt power supply.
If you have further questions contact 1-866-499-3966
Ext. 101. or visit www.tibscorp.com...