I got problems with a c-arm OEC 9400. After turned on the generator panel inticates kv ma time, and it is possible to turn x-ray on, but if you let it without press x-ray button during 10 seconds (aprox), appears an error KV IN ON ERROR then appears another: DATA CRC ERROR and the system turn halted
If somebody know about this problem I really appreciate your help, or if have a OEC9400 SERVICE MANUAL, please let me know. Thank you
We have the DATA CRC ERROR during the boot-up. I tried to cantact you but the e-mail adress that you sent is wrong:( Can you please contact me: dshatvoryan@yahoo.com
KV ON IN ERROR: KV was sensed without an X-ray requested. Fault with the HV tank. However your problem must be software loading onto the Technique Processor PCB which gave you DATA CRC ERROR. Solutions are as follows:
1. Boot floppy disk might be faulty, try a second or backup floppy disk, hopefully you have one.
2. Technique Processor PCB faulty. Replace it.
I have a perfect condition 9400 Schematic package 873920-03 rev P. Contact me if you want it. If you really need a 9400 service manual, I have a poor condition used one with notes. I have also a fair condition 9000 service manual, anyone wants it, please contact me. I have an entire set of good condition Uroview 2600 service documentation. I have a fair condition Uroview 2600 training manual with notes if anyone wants it.
Dear Mr. Yan
We need the service manual for OEC9400 so if you stil got this item please contact me. i am willing to buy it from you. Thanks.
Dion
email: dduran@medintek.com
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Manuel Esquer
OEC9400 DATA CRC ERROR
January 26, 2006 09:38
I got problems with a c-arm OEC 9400. After turned on the generator panel inticates kv ma time, and it is possible to turn x-ray on, but if you let it without press x-ray button during 10 seconds (aprox), appears an error KV IN ON ERROR then appears another: DATA CRC ERROR and the system turn halted
If somebody know about this problem I really appreciate your help, or if have a OEC9400 SERVICE MANUAL, please let me know. Thank you
to rate and post a comment
Manuel Casanova
Re: OEC9400 DATA CRC ERROR
January 27, 2006 11:29
please contac me at semmedica@gmail.com, i have a ful information over oec series
to rate and post a comment
Dvez
re: Re: OEC9400 DATA CRC ERROR
December 14, 2014 10:40
We have the DATA CRC ERROR during the boot-up. I tried to cantact you but the e-mail adress that you sent is wrong:( Can you please contact me: dshatvoryan@yahoo.com
to rate and post a comment
Edward Yan
Re: OEC9400 DATA CRC ERROR
January 27, 2006 11:35
KV ON IN ERROR: KV was sensed without an X-ray requested. Fault with the HV tank. However your problem must be software loading onto the Technique Processor PCB which gave you DATA CRC ERROR. Solutions are as follows:
1. Boot floppy disk might be faulty, try a second or backup floppy disk, hopefully you have one.
2. Technique Processor PCB faulty. Replace it.
I have a perfect condition 9400 Schematic package 873920-03 rev P. Contact me if you want it. If you really need a 9400 service manual, I have a poor condition used one with notes. I have also a fair condition 9000 service manual, anyone wants it, please contact me. I have an entire set of good condition Uroview 2600 service documentation. I have a fair condition Uroview 2600 training manual with notes if anyone wants it.
Edward Yan, B.S.E.E.
to rate and post a comment
Carlos Duran
Re: OEC9400 DATA CRC ERROR
March 08, 2006 07:51
hello Mr Yan:
please contact me about oec9400 needs.
thank you
to rate and post a comment
Edward Yan
Re: OEC9400 DATA CRC ERROR
March 09, 2006 08:37
Hello Mr. Duran:
How do you want me to contact you? Where are you located? What is your e-mail address?
My e-mail addresses are: edwardhyan@yahoo.com; edward.yan@va.gov
Edward Yan, B.S.E.E.
Certified Radiologic Electronics Specialist
to rate and post a comment
Dionicio Duran
Re: OEC9400 DATA CRC ERROR
December 15, 2009 10:54
Dear Mr. Yan
We need the service manual for OEC9400 so if you stil got this item please contact me. i am willing to buy it from you. Thanks.
Dion
email: dduran@medintek.com
to rate and post a comment
Paola Garcia
re: Re: OEC9400 DATA CRC ERROR
September 11, 2018 04:44
Could you give your email? i have a problem with the oec 9400
to rate and post a comment