sunil Kumar

Offset error -philips AVEU

October 18, 2005 12:48

I am getting offset error during scan and checks shows Bus line fail.Have changed I/O timer,TXS,TXR and FR-buff but no result.what else can be possible??

Log inor Register

to rate and post a comment

(57)
(1)
(15)

Barbaros Bulent Temiz

Re: Offset error -philips AVEU

November 13, 2005 07:41

Dear sunil,
check the 5v power supply just front of the Gantry.Also there is another power supply located back side of the Gantry.You can measure voltage but believe me manytimes we have seen this problem.

Bulent Temiz from BBTek