dear friend,
we had seen that problem several times.please check to the 5v power supply .sometimes ıt could come from it.you has to check specially receiver and tranceiver signal power supply.if you have always that problem ,it is very easy to fix that problem.
thanks for your reply.WE too have faced similar problem earlier and like you said it was TXR2 power supply.IN this case ,we have found that cable between gantry and console was cut and problem has been solved.strange may it seems but the bus line check was shown to be O.K.(we can't trust these diagnotic test of philips)
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
sunil Kumar
offset error OF01 philips SR7000
December 02, 2004 04:58
I am getting offset error OF01 during scan measurement all the diagnostic test are o.k.scanogram image is pretty o.k.any advice??
to rate and post a comment
Barbaros Bulent Temiz
Re: offset error OF01 philips SR7000
December 03, 2004 12:29
dear friend,
we had seen that problem several times.please check to the 5v power supply .sometimes ıt could come from it.you has to check specially receiver and tranceiver signal power supply.if you have always that problem ,it is very easy to fix that problem.
to rate and post a comment
sunil Kumar
Re: offset error OF01 philips SR7000
December 04, 2004 01:50
thanks for your reply.WE too have faced similar problem earlier and like you said it was TXR2 power supply.IN this case ,we have found that cable between gantry and console was cut and problem has been solved.strange may it seems but the bus line check was shown to be O.K.(we can't trust these diagnotic test of philips)
to rate and post a comment