marcelino diaz

olympus cv 100 display problem

April 30, 2009 02:15

olympus cv 100, The image is green, someone could help me?

Log inor Register

to rate and post a comment

David Till

Re: olympus cv 100 display problem

April 30, 2009 11:50

This problem could be related to a number of boards which are vertically mounted along the right side of the unit. Many times it is a number of leaky electrolytics which cause the problem. If you have any more questions you can email me at dave@lightsourcevideorepair.com

Log inor Register

to rate and post a comment

(7)
(2)
(3)

marcelino diaz

Re: olympus cv 100 display problem

May 01, 2009 10:46

do you think cleaning the boards could help or has to be change it? thanks