Sammy Arebalo
Only half an image seen probe
February 24, 2025 11:28
Only able to see a partial image when scanning the left of the image is good the right side of the image is not seen probe tested ok all tests passed . All three probe slots do the same thing any help?
to rate and post a comment
Stephane Lagace
re: Only half an image seen probe
February 25, 2025 10:55
Sammy, try to increase the prep to 1000ms or 2000ms, I know on some flat panel with single phase or old generator you need to increase the ms on the prep