Hi
I have a problem with Instrum OP200D Er 44 FIL
(First the unit doses not power On ,i check the Power supply board and the F3-2A was blow so the problem was at low voltage section i chavge the rectifire and the resistan.Now the unit can power On with the error 44 )
Any suggections will be helpfull

Log inor Register

to rate and post a comment

(11)
(3)

jagalindof

re: OP 200D

November 04, 2021 05:56

Hi Fanos.
Do you have the -25V and +25V (green LEDs in power supply board) on? Check that voltages are present in filament board and conections/cables to filament board from power supply, etc.