FANOS SPYROU

OP200D

September 20, 2018 08:51

I have a system that show Sy 30 PoH error code and the cassette holder does not move . Any suggestions? Looking for some service information or comment here...

Log inor Register

to rate and post a comment

(330)
(228)
(4)

Faustino Pereira

re: OP200D

September 21, 2018 10:42

The OP200D is a digital model and therefore there's no cassette holder for film. You might have an OP100 film type, or if it is an 200D, the sensor might have a vertical motion. Your error is on vertical movement of the rack. If the rack is stuck in the high position, remove X21 on the interface board and press down. The check limit switches or wiring. Link to access OP100 files sent trough Dotmed message center. Regards

Log inor Register

to rate and post a comment

(32)
(7)

Jose Antonio Galindo

re: re: OP200D

September 22, 2018 07:26

perfect explanation

Log inor Register

to rate and post a comment

(23)
(3)

FANOS SPYROU

re: re: OP200D

September 27, 2018 09:49

Hi Faustino
i do this with the X21 (Int Board) and the Cassette holder came on the original position, i check the signals H13 34V H14 15V H15 5V H1612VaC and its ok .
Now when i try give Xray i have the error SY28 POC

Log inor Register

to rate and post a comment

(32)
(7)

Jose Antonio Galindo

re: re: re: OP200D

September 27, 2018 02:16

It sounds that you have cable broken inside the cassette holder. This cable gives signals and supply to motors in cassette holder. I've changed most of them. You can put unit in test mode in order to try movements whitout x-ray and try to move cable inside. Sometimes this reveals some line broken.
Service manual helps to identify reference of cable and to test input/outputs of signals. It's easy to test.

Log inor Register

to rate and post a comment

(3)
(1)

Cory Stem

re: OP200D

September 21, 2018 10:42

Here is the troubleshooting manual-http://www.medteh.info/_fr/94/200_-1-.pdf

Log inor Register

to rate and post a comment

(19)
(4)

Zdenko Vukoja

re: OP200D

September 21, 2018 10:43

Microswitches. S
21 or S 22 may
not operate
properly or they hit
the mechanical
limit before
switching. Signals
RACKLIMSW and
RACKMIDSW
Control & power.
Rack motor or rack
motor drive in
Interface Board
may not operate.
Problem with
relay.
Use Sr 74 IOC.
Test the signals,
lift and lower the
rack. Adjust when
needed.
Check the control
from Core Module
to Interface board.
Check the motor
control red LED’s:
all should lit when
unit is rotating.
Check X16and
X17 connections.
No motor power. Check F2 on
Power Supply
board. Check
LED’s on Interface
board.
S 21, S22, SC4 or
C18, C19 in CR
models, X114,
C13, X6 Core
Module
Core Module, X 5,
Interface board, X
16 , C13, X114,
SC4or C18, M 5
Power Supply
board, X26, C5,
X15
200214 rev 3 Instrumentarium Dental 2

Log inor Register

to rate and post a comment

(32)
(7)

Jose Antonio Galindo

re: OP200D

September 21, 2018 10:43

Hi. But OP200D is digital (D indicate digital) and there is not cassette holder!
If not digital, first check leds in power supply. 35v must be lit. If not, check fuse and bridge rectifier (the small one)
If power supply is ok, Is the cassette holder stuck in upper position?

Log inor Register

to rate and post a comment

(32)
(7)

Jose Antonio Galindo

re: OP200D

September 21, 2018 10:44

Hi. But OP200D is digital (D indicate digital) and there is not cassette holder!
If not digital, first check leds in power supply. 35v must be lit. If not, check fuse and bridge rectifier (the small one)
If power supply is ok, Is the cassette holder stuck in upper position?

Log inor Register

to rate and post a comment