Opera T90 Receptor timeout

Tags: Radiology - R/F, GMM MERATE

(109)
(15)
(1)

Svilen Popov

Opera T90 Receptor timeout

December 22, 2016 04:53

Hello!
I have the following problem with GMM Opera T-90CEX;Generator:CPI Indico 100;TV:Infimed.
I haven/t fluoroscopy. All other functions works, but only fluoro doesn't. When I press foot switch,the generator shows error "Receptor timeout".

Log inor Register

to rate and post a comment

(62)
(14)

MM

re: Opera T90 Receptor timeout

December 27, 2016 10:17

Hello POPOV

Probably a fluoro contact or signal authorization is not coming; may be in table or infimed

1 person liked this
Log inor Register

to rate and post a comment

(109)
(15)
(1)

Svilen Popov

re: re: Opera T90 Receptor timeout

January 03, 2017 10:16

Yes maybe that is the problem, but how to find out where is the problem? In Opera, in Infimed, or in Indico...
I have the following bad situation: One rat ate some cables(some tv signal cables from Infimed GIM board to Indico TDI II board and cables from the console to Opera and Indico. But I change the tv signal cables with new ones and solder the interrupted cables from console...

Log inor Register

to rate and post a comment

(109)
(15)
(1)

Svilen Popov

re: re: Opera T90 Receptor timeout

January 03, 2017 10:16

And everything works perfect, except the fluoroscopy in SFD mode and whole DFR mode doesn't work. In Indico generator, on CPU board, lights only FLUORO_(inverted) led, witch mean only RAD mode...

Log inor Register

to rate and post a comment

(109)
(15)
(1)

Svilen Popov

re: re: Opera T90 Receptor timeout

January 03, 2017 10:16

In Opera, on the Interface board lights led for Manual collimator, like on 23S8 board lights the orange led(Manual), and I think that must light the greeone(ready).
I followed every cables, LEDS, relays.... and nothing

Log inor Register

to rate and post a comment

(109)
(15)
(1)

Svilen Popov

re: re: Opera T90 Receptor timeout

January 03, 2017 10:16

The other problem is that, I haven't nothing for Infimed GIM board(like schematics) and for Indico TDI II board...
Any other advice ? because I don't know where might be the problem....
Thank you in advance!!!

1 person liked this