I have the following problem with GMM Opera T-90CEX;Generator:CPI Indico 100;TV:Infimed.
I haven/t fluoroscopy. All other functions works, but only fluoro doesn't. When I press foot switch,the generator shows error "Receptor timeout".
Yes maybe that is the problem, but how to find out where is the problem? In Opera, in Infimed, or in Indico...
I have the following bad situation: One rat ate some cables(some tv signal cables from Infimed GIM board to Indico TDI II board and cables from the console to Opera and Indico. But I change the tv signal cables with new ones and solder the interrupted cables from console...
And everything works perfect, except the fluoroscopy in SFD mode and whole DFR mode doesn't work. In Indico generator, on CPU board, lights only FLUORO_(inverted) led, witch mean only RAD mode...
In Opera, on the Interface board lights led for Manual collimator, like on 23S8 board lights the orange led(Manual), and I think that must light the greeone(ready).
I followed every cables, LEDS, relays.... and nothing
The other problem is that, I haven't nothing for Infimed GIM board(like schematics) and for Indico TDI II board...
Any other advice ? because I don't know where might be the problem....
Thank you in advance!!!
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Svilen Popov
Opera T90 Receptor timeout
December 22, 2016 04:53
Hello!
I have the following problem with GMM Opera T-90CEX;Generator:CPI Indico 100;TV:Infimed.
I haven/t fluoroscopy. All other functions works, but only fluoro doesn't. When I press foot switch,the generator shows error "Receptor timeout".
to rate and post a comment
MM
re: Opera T90 Receptor timeout
December 27, 2016 10:17
Hello POPOV
Probably a fluoro contact or signal authorization is not coming; may be in table or infimed
to rate and post a comment
Svilen Popov
re: re: Opera T90 Receptor timeout
January 03, 2017 10:16
Yes maybe that is the problem, but how to find out where is the problem? In Opera, in Infimed, or in Indico...
I have the following bad situation: One rat ate some cables(some tv signal cables from Infimed GIM board to Indico TDI II board and cables from the console to Opera and Indico. But I change the tv signal cables with new ones and solder the interrupted cables from console...
to rate and post a comment
Svilen Popov
re: re: Opera T90 Receptor timeout
January 03, 2017 10:16
And everything works perfect, except the fluoroscopy in SFD mode and whole DFR mode doesn't work. In Indico generator, on CPU board, lights only FLUORO_(inverted) led, witch mean only RAD mode...
to rate and post a comment
Svilen Popov
re: re: Opera T90 Receptor timeout
January 03, 2017 10:16
In Opera, on the Interface board lights led for Manual collimator, like on 23S8 board lights the orange led(Manual), and I think that must light the greeone(ready).
I followed every cables, LEDS, relays.... and nothing
to rate and post a comment
Svilen Popov
re: re: Opera T90 Receptor timeout
January 03, 2017 10:16
The other problem is that, I haven't nothing for Infimed GIM board(like schematics) and for Indico TDI II board...
Any other advice ? because I don't know where might be the problem....
Thank you in advance!!!