Over 500 Total Lots Up For Auction at One Location - NJ 06/15

OR 35 MAGNET QUESTION

Selcuk Akin

OR 35 MAGNET QUESTION

July 02, 2007 11:05

FOR QUENCH OR 35 MAGNET WHEN İNSTALLİNG DSL PROBE DO WE READ A SHORT CIRCUIT?

NOTE: ERROR CODE 105 IS READ N THE DİSPLAY AT THİS TİME

Log inor Register

to rate and post a comment

(75)
(2)
(2)

Iosif Sogolov

Re: OR 35 MAGNET QUESTION

July 03, 2007 07:29

is the switch heater on?

Log inor Register

to rate and post a comment

(42)
(1)
(17)

Marc Fessler

Re: OR 35 MAGNET QUESTION

July 03, 2007 09:29

You probably know by now, but you should read less then 1 ohm.
Only do this on magnet without field.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Therapeutic Technologies Surplus Auctions June 15th

Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)