PALI System doesn't work

Tags: MRI, SIEMENS

(11)
(2)

Andrey Belousov

PALI System doesn't work

December 11, 2012 10:52

We have a problem with Siemens Magnetom Harmony.
PALI system doesn't work (doesn't load).
After starting diagnostic tests system say us about two problems:
1. MSUP system: Supervision 40 V failure
2. PTAB system: I2C Power supply failure

What it can be?

Log inor Register

to rate and post a comment

(62)
(11)
(14)

James Fower

re: PALI System doesn't work

December 11, 2012 06:27

We just had same problem. Need to run tests on communications. More likely will find a bad FO cable.

Log inor Register

to rate and post a comment

(11)
(2)

Andrey Belousov

re: re: PALI System doesn't work

December 11, 2012 07:02

After Communication test we have:

1. MSUP Status: MSUK2200
- Supervision 40 V = fault
2. PTAB Status: PTABK2200
- I2C Power supply = fault
- EEPROM state = fault

Log inor Register

to rate and post a comment

(20)
(5)
(1)

Gandhi Mendez

re: re: re: PALI System doesn't work

December 12, 2012 10:23

Look at FUN manual that usually comes with system.
Are you reinstalling it?
As Fower says check FO and cables, connectors, blended pins.
Usually module/system previous (in communication chain) than reported is really broken instead of repoted, In your cases, check system/module that is before MSUP or Ptab in chain comunication.

Log inor Register

to rate and post a comment