We have the following problem with an old generator CGR Phasix 65.
We obtain the error c24 (parity check error).
Would anyone on this forum have any idea on how this problem can be solved ?
The generator will eventually be replaced january 2016.
Many thanks,
R. Vanclooster
Dear Umberto,
I thank you for your reply ; the error occurred after switch on.
I was lucky to receive a surplus Eprom1 card from a belgian dealer ; the problem then disappeared.
Careful study showed that there was a bad contact between the eprom1 card and the chassis.
Again my sincere thanks for your reply.
Roger Vanclooster.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Roger Vanclooster
Phasix 65
October 12, 2015 03:48
We have the following problem with an old generator CGR Phasix 65.
We obtain the error c24 (parity check error).
Would anyone on this forum have any idea on how this problem can be solved ?
The generator will eventually be replaced january 2016.
Many thanks,
R. Vanclooster
to rate and post a comment
umberto sciunnach
re: Phasix 65
October 12, 2015 06:27
Hello
You have this problem after switch on or during pre/xray?
Regards
Umberto
to rate and post a comment
Roger Vanclooster
re: re: Phasix 65
October 27, 2015 08:25
Dear Umberto,
I thank you for your reply ; the error occurred after switch on.
I was lucky to receive a surplus Eprom1 card from a belgian dealer ; the problem then disappeared.
Careful study showed that there was a bad contact between the eprom1 card and the chassis.
Again my sincere thanks for your reply.
Roger Vanclooster.
to rate and post a comment