Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Philipis BV 300 plus

Tags: C-Arm

(6)

kcmservice

Philipis BV 300 plus

September 11, 2018 09:59

Good Day Servicemen, Im in Need of Service Manual for Philips BV 300 Plus.

Log inor Register

to rate and post a comment

DMEDS

re: Philipis BV 300 plus

September 16, 2018 02:48

i have the manuals for BV 300 ,
if you need it contact ,
magmesme@gmail.com
regards

Log inor Register

to rate and post a comment

(332)
(37)
(60)

Gary Benitez

re: re: Philipis BV 300 plus

February 19, 2024 12:43

Do you have a battery charger for a Pulsera?

Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: Philipis BV 300 plus

September 16, 2018 02:49

Hi,
If you have a problem, then write your problem and then we are looking to this.

A manual alone make most not happy. Then when you have a Philips training, then you have a manual ...

Log inor Register

to rate and post a comment

(6)

kcmservice

re: re: Philipis BV 300 plus

September 17, 2018 10:30

Hi, it shows on the Screen " Built in Self test Failed",
But still I can expose.

Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: re: re: Philipis BV 300 plus

September 18, 2018 08:59

Hi,

With "Screen" you mean the stand display ? maybe yes.

Ok, then you have maybe the Stand Error 010 on the system.
This Error 10 say, that the timebase by the Stand (C-Arm Part) are not correct. This is most the point of a empty batterie on the Standcontroller.

Have you look on this ? Have you also look to the logfile?

You can connect by VT100 by the standcontroller. That is the board on the left side - SHA01. Then you have a lot of information and you must not have any Philips software.

Kind regards
Willi

Log inor Register

to rate and post a comment

(6)

kcmservice

re: re: re: re: Philipis BV 300 plus

September 20, 2018 08:57

hi sir,
I haven't look at the logfile, I dont have access to that, Thats why Im looking for manuals and or software utility tools.

Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: re: re: re: re: Philipis BV 300 plus

September 20, 2018 04:51

Hi,

The problem is, that the software X-Scope is protected and you can nothing do with it.

But, connect your PC with the VT 100 Mode to the seriell connector and you will see the life log.

Then you see the Error normaly.

Kind regard
Willi

Log inor Register

to rate and post a comment

(6)

kcmservice

re: re: re: re: re: re: Philipis BV 300 plus

September 22, 2018 07:38

Hi,

The error code was "M 014"
Built in selftest failed.
But I can still use the machine.

thank you.

Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: re: re: re: re: re: re: Philipis BV 300 plus

September 23, 2018 07:35

Hi,

Now Error M014 = SUCO-COMBO Board self test Error due do failure of the time chip. Replace the SHA1 Board.

But look, when you to this, the software must be reloaded.

Kind regards
Willi

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)