Philips ATL HDI 5000. Error 0021.

October 17, 2011 10:08

Good day, gentlemen! I Served Ultrasound Philips ATL HDI 5000. Anyone encountered error 0021? In what may be the problem? (Power supply and the computer are OK)
Thanks to all !

Log inor Register

to rate and post a comment

Jim Carr

Re: Philips ATL HDI 5000. Error 0021.

October 17, 2011 05:09

Error banner 0021 indicates a task crash. Check the error log to see if there are other errors, and run the Comprehensive tests if available. Otherwise, no action is indicated.
Jim Carr
Unisyn Medical

Log inor Register

to rate and post a comment

Mohd Dani Sanin

Re: Philips ATL HDI 5000. Error 0021.

October 18, 2011 09:51

Here I paste the error detail. there is no recommendation stated.

0018 VDiagsSoftware.c
This is only a test message. It occurs when VDiagsSftw_PriorityAlert is executed to
verify a banner can be generated.

0019 VPMPowerMonitor.c
The main task was suspended or crashed at the time the standby switch was set to off.
No recommendations.

0020 VPMClockMonitor.c
Monitor task crash. No recommendations.

0021 VPMClockMonitor.c
Task crash. No recommendations.

0022 mainVxworks.c
Chkdsk failed with errors that it could not fix. The most likely cause is a failed hard

Log inor Register

to rate and post a comment

(19)

Pablo Ceretani

Re: Philips ATL HDI 5000. Error 0021.

October 27, 2011 10:13

Boot with Optical Disk,and probe the hard soft. It´s work, replace hard drive or reload software.

Log inor Register

to rate and post a comment

Hugo Scharovsky

Re: Philips ATL HDI 5000. Error 0021.

October 31, 2011 10:07

привет Mikhail

A good friend once told me the following:

In systems with software revision lower than 200.xx, the aleatory occurence of the blue banner message 0021, displayed during system power down on systems with one or more NetLink devices enabled, is caused by suspended tasks and they are absolutely benigns.
You can eliminate it through a software upgrade, but it is not mandatory, and on the other hand is expensive.You Should ask to your customer about how to proceed.
Regards
Ricardo Scacchi

Log inor Register

to rate and post a comment

(56)
(19)

Mikhail Zakurdaew

Re: Philips ATL HDI 5000. Error 0021.

November 01, 2011 04:16

Good day! Thank you very much for response and information! Problem solved by replacing CPU system.

Log inor Register

to rate and post a comment