PHILIPS BV 300 PLUS C-ARM ERROR CODES

Yusuf Odabas

PHILIPS BV 300 PLUS C-ARM ERROR CODES

December 03, 2012 01:09

I WOULD LIKE TO KNOW BEST ADVICE FOR THESE ERROR CODES FROM OUR BV 300 PLUS C-ARM UNIT'S.

M341
M385

BEST DESCRIPTION AND RESULT...
PLEASE SEND ME YOUR VALUABLE ADVICE. I'D LIKE TO CLEAR SOONEST I CAN .

1 person liked this
Log inor Register

to rate and post a comment

(16)
(3)
(1)

charles hauer

re: PHILIPS BV 300 PLUS C-ARM ERROR CODES

December 03, 2012 11:34


Erro M385 END OF BATTERY LIFE
First, see if the battery charger is ok
Second, if the battery charger is ok, change the batteries, all batteries!

Erro M341 can be caused by errors of M385.





Erro M385 END OF BATTERY LIFE
First, see if the battery charger is ok
Second, if the battery charger is ok, change the batteries, all batteries!

Erro M341 can be caused by errors of M385.




Erro M385 END OF BATTERY LIFE
First, see if the battery charger is ok
Second, if the battery charger is ok, change the batteries, all batteries!

Erro M341 can be caused by errors of M385.




Erro M385 END OF BATTERY LIFE
First, see if the battery charger is ok
Second, if the battery charger is ok, change the batteries, all batteries!

Erro M341 can be caused by errors of M385.






Log inor Register

to rate and post a comment

(36)
(7)
(1)

Jim Shami

re: re: PHILIPS BV 300 PLUS C-ARM ERROR CODES

December 18, 2012 11:06

if serice manula available please send a copy or if its hard copy i can arrange to copy and resend please advice
thanks

Log inor Register

to rate and post a comment

(15)
(5)
(1)

Hamdy Abou Abdou

re: re: re: PHILIPS BV 300 PLUS C-ARM ERROR CODES

September 27, 2013 10:10

Dear Sir


we have some problem in Philips BV 300 C-arm


we looking for service manual for Philips BV 300 C-arm

or solution for the following error (sup-system error M208)

this error come if you want to imaging the back or abdomen

Log inor Register

to rate and post a comment

(15)
(5)
(1)

Hamdy Abou Abdou

re: re: PHILIPS BV 300 PLUS C-ARM ERROR CODES

September 27, 2013 10:09

Dear Sir


we have some problem in Philips BV 300 C-arm


we looking for service manual for Philips BV 300 C-arm

or solution for the following error (sup-system error M208)

this error come if you want to imaging the back or abdomen

Log inor Register

to rate and post a comment

(15)
(5)
(1)

Hamdy Abou Abdou

re: re: PHILIPS BV 300 PLUS C-ARM ERROR CODES

September 27, 2013 10:10

Dear Sir


we have some problem in Philips BV 300 C-arm


we looking for service manual for Philips BV 300 C-arm

or solution for the following error (sup-system error M208)

this error come if you want to imaging the back or abdomen

please send me on hamdymaa@gmail.com

Log inor Register

to rate and post a comment

(2)

Miguel Torres

re: re: re: PHILIPS BV 300 PLUS C-ARM ERROR CODES

October 04, 2018 09:02

Hola! I have this error M 208 , when kv are up 60 kv in manual or Automatic fluoride

Log inor Register

to rate and post a comment

(15)
(5)
(1)

Hamdy Abou Abdou

re: PHILIPS BV 300 PLUS C-ARM ERROR CODES

September 27, 2013 10:09

Dear Sir


we have some problem in Philips BV 300 C-arm


we looking for service manual for Philips BV 300 C-arm

or solution for the following error (sup-system error M208)

this error come if you want to imaging the back or abdomen

2 people liked this
Log inor Register

to rate and post a comment

(26)
(3)

Chuck

re: re: PHILIPS BV 300 PLUS C-ARM ERROR CODES

March 14, 2022 10:14

Good day.
Have you managed to get the manual for the Endura BV300?
I have the same error now, M208!
Please help!
Thanks.

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: re: re: PHILIPS BV 300 PLUS C-ARM ERROR CODES

March 15, 2022 11:39

XACT_GONE
M 208
Subsystem error, system still available
It‘s possible to continue

Unexpected X-ray switch off
Check board SHA4 and the generator unit SE. Check cabling.



Log inor Register

to rate and post a comment

(7)

rockbob

re: re: PHILIPS BV 300 PLUS C-ARM ERROR CODES

March 15, 2022 11:37

check the SHA4 board and cabling. try re-seating the board

Log inor Register

to rate and post a comment