Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Philips BV Pulsera IP's

Tags: C-Arm, PHILIPS

(22)
(10)

Brian

Philips BV Pulsera IP's

May 06, 2013 01:27

I need help changing the ip addresses on a Philips BV Pulsera c-arm. Can anyone help me with that?

1 person liked this
Log inor Register

to rate and post a comment

(5)

Thomas Williams

re: Philips BV Pulsera IP's

May 07, 2013 08:49

Go to the patient screnn then the setup screen on the workstation. You will see the IP address. The 192.XXX.XXX.XX address is for service.

Log inor Register

to rate and post a comment

(22)
(10)

Brian

re: re: Philips BV Pulsera IP's

May 07, 2013 12:44

Hello.
I was following the instructions in the service manual, saw the part you described, pressed the "enable service" button, and the next step is to "open BV Scope." This is the part where I get hung up. Is this a service software tools of some sort?
Thanks.

Log inor Register

to rate and post a comment

(5)

Thomas Williams

re: re: re: Philips BV Pulsera IP's

May 08, 2013 10:44

What are you wanting to change? What version BV Pulsera do you have?

Log inor Register

to rate and post a comment

(22)
(10)

Brian

re: re: re: re: Philips BV Pulsera IP's

May 08, 2013 10:48

Hello. I need to change the worklist host table. Not sure which version. How can I tell?

Log inor Register

to rate and post a comment

(2)

oscar de la rosa

re: re: re: re: re: Philips BV Pulsera IP's

June 27, 2014 05:05

wich is your mail?, you have to connect a PC and enter a a IP addressContact me j.elizalde@vensi.com.mx