mostly, you only need replace the II cascade generator for the II tube.Pls check the Red led on II cascade generator, also check the power supply for II cascade generator.
hello
the led H1 in II casscad is on , we check the power to this part it is normal -15v & +15v
this error appear directly after we start the machine .
hello
thank you for your information , we remove this part and clean it from silicon , after thatr we connect it with system , H1 led is off now and there no error on machine , but still there no image as no high voltage come from this part .
regards
if you reinstall the same HVPS, with new isolating silicone, and not error and not red led.try with fluoroscopy, the KV indicator should not go to maximum 120KV. try puting an open chasis to see if you have x-ray.
hello
the red led on this part on , and machine give x-ray and appear image but not good quility image .
if you must change this part , are you have and how much coast?
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Abdelmuin Khatib
philips bv25
February 09, 2012 08:16
i have c-arm model philips bv25 , this system give me error 1 .
so it need to change HV-image intsfire , i hope i can find it .
to rate and post a comment
Steven Zhao
Re: philips bv25
February 09, 2012 02:05
mostly, you only need replace the II cascade generator for the II tube.Pls check the Red led on II cascade generator, also check the power supply for II cascade generator.
to rate and post a comment
Henrique Gomes
re: Re: philips bv25
June 21, 2016 09:06
Hi Mr Steven,can u e-mail me (gomesh@gmx.com) ? (BV25 issue), tks
to rate and post a comment
Henrique Gomes
re: Re: philips bv25
June 21, 2016 09:06
Hello, can you e-mail me (neutrinno@zipmail.com) or inform yours ? (BV25 issue)
to rate and post a comment
Jorge Cid Alvarez
Re: philips bv25
February 10, 2012 09:15
Yes, II, almost never fails.
to rate and post a comment
Abdelmuin Khatib
Re: philips bv25
February 12, 2012 11:29
hello
the led H1 in II casscad is on , we check the power to this part it is normal -15v & +15v
this error appear directly after we start the machine .
to rate and post a comment
Jorge Cid Alvarez
Re: philips bv25
February 13, 2012 12:12
That part HVPS 980714020301 is very failing (normal price about usd 2000) I got a couple of them.
HVPS has a silicon green (red or gray) rubber, which become liquid and lost his isolating characteristic.
red led is failing indication.
sound like cli-cli-cli... is typical fail condition.
voltage incoming to HVPS... normaly are ok.
sometimes is one cable who is broken...and it is diffical to repair, because, isolating sail is deffical to obtain by suppliers.
you must disconect from the II, wiht metalic silver connector included as part of HVPS.
to rate and post a comment
ikram baig
Re: philips bv25
February 14, 2012 12:14
II supply should be replaced.
to rate and post a comment
Abdelmuin Khatib
Re: philips bv25
February 20, 2012 10:11
hello
thank you for your information , we remove this part and clean it from silicon , after thatr we connect it with system , H1 led is off now and there no error on machine , but still there no image as no high voltage come from this part .
regards
to rate and post a comment
Jorge Cid Alvarez
Re: philips bv25
February 22, 2012 10:13
if you reinstall the same HVPS, with new isolating silicone, and not error and not red led.try with fluoroscopy, the KV indicator should not go to maximum 120KV. try puting an open chasis to see if you have x-ray.
to rate and post a comment
Abdelmuin Khatib
Re: philips bv25
February 12, 2012 11:31
hello
the red led on this part on , and machine give x-ray and appear image but not good quility image .
if you must change this part , are you have and how much coast?
to rate and post a comment
Tibor Ivanszky
Re: philips bv25
February 11, 2012 11:46
Is the error 1 appears immediately on the display or after you press the hidden button?
to rate and post a comment