philips bv25

(59)
(5)
(4)

Abdelmuin Khatib

philips bv25

February 09, 2012 08:16

i have c-arm model philips bv25 , this system give me error 1 .
so it need to change HV-image intsfire , i hope i can find it .

2 people liked this
Log inor Register

to rate and post a comment

(16)
(2)

Steven Zhao

Re: philips bv25

February 09, 2012 02:05

mostly, you only need replace the II cascade generator for the II tube.Pls check the Red led on II cascade generator, also check the power supply for II cascade generator.

Log inor Register

to rate and post a comment

(120)
(15)
(15)

Henrique Gomes

re: Re: philips bv25

June 21, 2016 09:06

Hi Mr Steven,can u e-mail me (gomesh@gmx.com) ? (BV25 issue), tks

Log inor Register

to rate and post a comment

(120)
(15)
(15)

Henrique Gomes

re: Re: philips bv25

June 21, 2016 09:06

Hello, can you e-mail me (neutrinno@zipmail.com) or inform yours ? (BV25 issue)

Log inor Register

to rate and post a comment

(59)
(9)
(1)

Jorge Cid Alvarez

Re: philips bv25

February 10, 2012 09:15

Yes, II, almost never fails.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

Re: philips bv25

February 12, 2012 11:29

hello
the led H1 in II casscad is on , we check the power to this part it is normal -15v & +15v
this error appear directly after we start the machine .

Log inor Register

to rate and post a comment

(59)
(9)
(1)

Jorge Cid Alvarez

Re: philips bv25

February 13, 2012 12:12

That part HVPS 980714020301 is very failing (normal price about usd 2000) I got a couple of them.

HVPS has a silicon green (red or gray) rubber, which become liquid and lost his isolating characteristic.

red led is failing indication.
sound like cli-cli-cli... is typical fail condition.

voltage incoming to HVPS... normaly are ok.

sometimes is one cable who is broken...and it is diffical to repair, because, isolating sail is deffical to obtain by suppliers.


you must disconect from the II, wiht metalic silver connector included as part of HVPS.

Log inor Register

to rate and post a comment

(5)

(15)

ikram baig

Re: philips bv25

February 14, 2012 12:14

II supply should be replaced.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

Re: philips bv25

February 20, 2012 10:11

hello
thank you for your information , we remove this part and clean it from silicon , after thatr we connect it with system , H1 led is off now and there no error on machine , but still there no image as no high voltage come from this part .
regards

Log inor Register

to rate and post a comment

(59)
(9)
(1)

Jorge Cid Alvarez

Re: philips bv25

February 22, 2012 10:13

if you reinstall the same HVPS, with new isolating silicone, and not error and not red led.try with fluoroscopy, the KV indicator should not go to maximum 120KV. try puting an open chasis to see if you have x-ray.

Log inor Register

to rate and post a comment

(59)
(5)
(4)

Abdelmuin Khatib

Re: philips bv25

February 12, 2012 11:31

hello
the red led on this part on , and machine give x-ray and appear image but not good quility image .
if you must change this part , are you have and how much coast?

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

Re: philips bv25

February 11, 2012 11:46

Is the error 1 appears immediately on the display or after you press the hidden button?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS Mx-8000 IDT/Brilliance 16 Slice CT Scanner -----Worldwide leader!!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)