I have a BV29 which when switched on displays 'error 4'.
The monitors are both blank and no exposures are possible.
The SE15 board has a red LED on MMPNR.
Check to see if the memory box below the monitors lights up if it does not then the problem is the power stage the smps power supply inside the memory box is defective happens quite often, you will need to replace this
Paul Baker
Philips BV29 - error 4
December 07, 2011 09:19
I have a BV29 which when switched on displays 'error 4'.
The monitors are both blank and no exposures are possible.
The SE15 board has a red LED on MMPNR.
Can anyone help?
to rate and post a comment
Ahmed Badr
Re: Philips BV29 - error 4
December 07, 2011 05:43
This error is from the hard disk either software or the hard disk itself
to rate and post a comment
Felix Donoghue
Re: Philips BV29 - error 4
December 08, 2011 12:24
Check to see if the memory box below the monitors lights up if it does not then the problem is the power stage the smps power supply inside the memory box is defective happens quite often, you will need to replace this
to rate and post a comment
BRUNO GOMES
re: Philips BV29 - error 4
February 10, 2020 09:25
Hello Paul,
Did you manage to solve this problem ? I have the same problem with my C-Arm BV29. Can you Help me ?
Thank you in advace.
to rate and post a comment