Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

PHILIPS BV29 ERROR

Posted by an Anonymous User

PHILIPS BV29 ERROR

April 14, 2014 08:53

DEAR FRIENDS
I am Getting ERROR-1 in PHILIPS BV 29 C-ARM
ALSO FOUND RED LIGHT ON TVAL ON BOARD SE15
ANY HELP OR IDEA

Log inor Register

to rate and post a comment

(40)
(7)

Roberto Cruz

re: PHILIPS BV29 ERROR

April 14, 2014 07:02

Yes error 1 not have htv your verific power supply

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: PHILIPS BV29 ERROR

April 15, 2014 07:55

It is a TV alarm, the H3 LED 1 is activated by a synchron detector on the S15. A VSYNC-L signal is missing there or the signal is high which is coming from the TV trolley through the large cable. Check the pins "M" and conductivity of "M" between the connectors. Unfortunately, my manual is not following this signal within the Image Processor. Try to find where the "M" is going. Also, there are indicators within the Image Processing Unit. Let me know them.

Log inor Register

to rate and post a comment

(15)
(4)

DEVENDRAN DRAN

re: PHILIPS BV29 ERROR

April 15, 2014 10:17

This error indicates CCTV camera faulty or video signal is 0 it must be .8 to 1v dc

1 person liked this
Log inor Register

to rate and post a comment

(40)
(6)
(1)

Santosh Kumar Tiwari

re: re: PHILIPS BV29 ERROR

April 21, 2014 11:46

Thanks For helping me.
i Will Check the CCTV Camera and inform you the update.

Log inor Register

to rate and post a comment

(40)
(6)
(1)

Santosh Kumar Tiwari

re: re: PHILIPS BV29 ERROR

April 21, 2014 11:46

Thanks Devendran
I Will Check the Video Signal.
If you know the test point please let me know.
Because at Video Cable there is 00 DC Voltage.

Log inor Register

to rate and post a comment

(177)
(20)
(3)

Chris Baklik

re: PHILIPS BV29 ERROR

April 15, 2014 03:49

I have the service manual email me crisb1955@aol.com

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: PHILIPS BV29 ERROR

April 16, 2014 08:36

I found the signal, on SE 17 there is a synchron separator from VIBS signal. This is coming from the Camera. Close to the camera there is BA1 board, MP3 testpoint is for the VIBS signal.

Log inor Register

to rate and post a comment

(40)
(6)
(1)

Santosh Kumar Tiwari

re: re: PHILIPS BV29 ERROR

April 21, 2014 11:46

i Checked SE17 in another working BV29 system and this all SE Boards found OK

Log inor Register

to rate and post a comment

Mohammed Abdullah

re: PHILIPS BV29 ERROR

April 16, 2014 03:38

i have the manuals if you need it contact me

magmesme@gmail.com

regards

Log inor Register

to rate and post a comment

(40)
(6)
(1)

Santosh Kumar Tiwari

re: re: PHILIPS BV29 ERROR

April 21, 2014 11:46

My Mail ID in trivenibiotronix@yahoo.co.in Pl Mail me the Service manual.