Please I need your help, my C Arm PHILIPS BV300 shows error mesage M205, This error, don“t show in the manual service, somebody knows what is this?
Thanks in advance.
M 205 TIMER
System not available Switch system off/on Call service if problem is persistent
The BIST of the Pulse master board has failed because one of the timers could not be programmed. Replace pulse master board SHA3.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Juan Carlos Velasco
Philips BV300 error M205
April 24, 2015 08:20
Please I need your help, my C Arm PHILIPS BV300 shows error mesage M205, This error, don“t show in the manual service, somebody knows what is this?
Thanks in advance.
to rate and post a comment
Lenny Place
re: Philips BV300 error M205
April 26, 2015 11:50
M 205 TIMER
System not available Switch system off/on Call service if problem is persistent
The BIST of the Pulse master board has failed because one of the timers could not be programmed. Replace pulse master board SHA3.
to rate and post a comment
DMEDS
re: Philips BV300 error M205
May 08, 2015 07:13