Over 500 Total Lots Up For Auction at One Location - NJ 06/15

PHILIPS C ARM BV300 IMAGE PROBLEM

Tags: C-Arm, PHILIPS

(53)
(9)

EFREN FRANCO

PHILIPS C ARM BV300 IMAGE PROBLEM

January 05, 2016 04:29

Hello: I have a problem with the Image Intensifier in a BV 300 it only shows a bright circle and I´m looking for the technical information about the I. I. to check references voltages and signals. Thanks for the comments you can do.

1 person liked this
Log inor Register

to rate and post a comment

(138)
(18)
(1)

umberto sciunnach

re: PHILIPS C ARM BV300 IMAGE PROBLEM

January 06, 2016 08:31

Hello
Witch kindle od II you have ?
5/9 or 5/9/14 inch.
You habe different supply for two ib.
Regards
Umberto
Umbertosc1957@gmail.com

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: PHILIPS C ARM BV300 IMAGE PROBLEM

January 06, 2016 08:31

Remove camera and see if the output screen of the II.glows during fluoro.

Log inor Register

to rate and post a comment

(53)
(9)

EFREN FRANCO

re: re: PHILIPS C ARM BV300 IMAGE PROBLEM

January 06, 2016 04:58

The I I has the 5, 7, and 9" fields. The image glows in the output of the II.

Log inor Register

to rate and post a comment

(138)
(18)
(1)

umberto sciunnach

re: re: re: PHILIPS C ARM BV300 IMAGE PROBLEM

January 06, 2016 05:41

Sure is very difficoult tath the II is brocken.
Witct kindle of tv sistem you have?
Regards
Umberto

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: re: re: PHILIPS C ARM BV300 IMAGE PROBLEM

January 07, 2016 09:08

Under the monitor(s) there is a multi output power supply for the imaging system, check this out. tibor@ivanszky.com

Log inor Register

to rate and post a comment

(53)
(9)

EFREN FRANCO

re: re: re: re: PHILIPS C ARM BV300 IMAGE PROBLEM

January 08, 2016 08:37

Thanks Tibor I'll check this Power supply,

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: re: re: re: re: PHILIPS C ARM BV300 IMAGE PROBLEM

January 08, 2016 11:30

use my email for further communication if necessary tibor@ivanszky.com
If you see image on the output screen you have to go further troubleshooting on other imaging parts. I have a working system in my workshop, though with a larger II. Maybe some parts are similar with yours.
Rgds, Tibor

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: re: re: re: re: re: PHILIPS C ARM BV300 IMAGE PROBLEM

January 08, 2016 02:05

To be precise, if you put an object on the II surface the image you must see on the output of the II must be a sharp minimized equivalent of the object in all magnification. If your case is this your II and power supply are not damaged. There could be unsharpness on different magnifications which you may adjust. Let me know your result.
tibor@ivanszky.com

Log inor Register

to rate and post a comment

(53)
(9)

EFREN FRANCO

re: re: re: re: re: re: re: PHILIPS C ARM BV300 IMAGE PROBLEM

January 09, 2016 06:59

]I will contact you via email.Thank you.

Log inor Register

to rate and post a comment

(138)
(18)
(1)

umberto sciunnach

re: PHILIPS C ARM BV300 IMAGE PROBLEM

January 07, 2016 06:07

Hello
You must decide witch engenier follow.
Ear his instruction and follow him.
You will never solve the problem.
Regards
Umberto

Log inor Register