Philips DuoDiagnost error 3232002

(15)
(3)

Shawn Lincoln

Philips DuoDiagnost error 3232002

February 14, 2013 08:26

Philips DuoDiagnost error 3232002

If anyone has details of this error related to the table bucky, please send info to SLNetMed@gmail.com

1 person liked this
Log inor Register

to rate and post a comment

(16)
(9)

tanya Lobyam

re: Philips DuoDiagnost error 3232002

February 18, 2013 08:10

Hi Error 3232002 mean, GEO section, LM motor fail to start

Log inor Register

to rate and post a comment

(45)
(8)

carlos enrique salazar gea

re: re: Philips DuoDiagnost error 3232002

August 13, 2014 08:09

Could you tell me what does the errors 3116001,1201008 and 3232001 means?

Log inor Register

to rate and post a comment

(47)
(4)

Milos Krstic

re: Philips DuoDiagnost error 3232002

March 01, 2013 09:55

Hi, you have problem with serial changer - Left Motor of cassette holder is defective or mostly flat cable is broken. Also could be ser.changer ctrl. pcb D80 defective. You have check if something is inside of serial changer (some cassette marks, ...).

1 person liked this
Log inor Register

to rate and post a comment

Marian Cristea

re: re: Philips DuoDiagnost error 3232002

February 11, 2016 11:50

Hi,
What could cause the errorr "Cassette tray not closed"?It appears randomly not everi day.
Thanks

Log inor Register

to rate and post a comment

(12)
(1)

GABRIEL MENESES

re: Philips DuoDiagnost error 3232002

March 07, 2013 02:53

Hi Mr Shawn, your problem is in the serial changer.
Is clean iin the left motor?
The left motor is hot? Or the encoder is loosing? Exchange the left motor
The flat cable is disconected ? Check both sides in the flat cable
Replace flat cable

Log inor Register

to rate and post a comment