Philips DuoDignost erro 3231003

Posted by an Anonymous User

Philips DuoDignost erro 3231003

January 01, 2016 10:26


Gentelmen ,

Error 3231003 is displayed .
When freecasate technique is selected , colimator size not controlled ; x ray image is 5cm width .

Please details & steps to solve the problem

Regards

1 person liked this
Log inor Register

to rate and post a comment

(29)
(2)

Fastone Oscar RUGOTWA

re: Philips DuoDignost erro 3231003

January 02, 2016 02:35

Hi my friend
This 3231003 is a fatal error
Description:
GEO:SEC:Mech.or electr.error:RM Motor moved too.
Possiple cause:
Belt broken,check the drive
Regards Fastone Oscar
CEO FANESTECH (T) LTD.

Log inor Register

to rate and post a comment

(29)
(2)

Fastone Oscar RUGOTWA

re: Philips DuoDignost erro 3231003

January 02, 2016 02:36

Hi,
Where are you from, am engineer Fastone from Tanzania,are you a technical personel or user?
If you need an Eng for rectifying the problem i can help my friend.
However if you are a technician i can guide what to do for the machine.
Regards Eng.Fastone Oscar.

Log inor Register

to rate and post a comment

(23)
(4)

necdet sahin

re: Philips DuoDignost erro 3231003

January 02, 2016 02:37

hello

Write the e-mail.
necdet şahin
necexray@hotmail.com

Log inor Register

to rate and post a comment

(67)
(7)

Galal

re: re: Philips DuoDignost erro 3231003

January 03, 2016 10:19

thanks alot

Log inor Register

to rate and post a comment

(138)
(18)
(1)

umberto sciunnach

re: Philips DuoDignost erro 3231003

January 02, 2016 02:39

Hello
On service manual there is written the type of error.
If I remember is security error and you can found simply.
On the base of remote control there is a plastic box.
Read error on the manual and I can help you how eliminate.
My mail umbertosc1957@gmail.com
My mobile for whatsapp +393891529020.
Regards
Umberto

Log inor Register

to rate and post a comment

(67)
(7)

Galal

re: re: Philips DuoDignost erro 3231003

January 03, 2016 10:19

thanks a lot

Log inor Register

to rate and post a comment

(29)
(2)

Fastone Oscar RUGOTWA

re: re: re: Philips DuoDignost erro 3231003

January 04, 2016 10:09

Hi
As i know about this error,you can connect pc kit on the console and through you pc if x scope if istalled,you can open and go to error log,and check the possible cause.
If not open the serial changer on the table and move out the whole serial changer,then check the Belt for Right motor and fix it,the machine ill work.
My mail :fastone.oscar@yahoo.com
Regards Fastone

Log inor Register

to rate and post a comment

Mostafa Lotfy

re: Philips DuoDignost erro 3231003

May 20, 2016 04:36

Check RM Encoder disc and adjust with LM Encoder disc , to be with in range as visual.
But the high possibility "flat cable" inside SCE.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)