Philips Easy Diagnost gives a system error when I try to screen. Otherwise the x-ray works perfect, only screening gives error.
If there is a Philips technician with a service doggle for both Generator and DSI cabinets willing to fly to Kenya I am willing to discuss.
The xray works but the fluoro gives system error on the scopo user interface...I believe it has something to do with the DSI cabinet. It was working then I was setting up optical disk in the DSI cabinet and the fluoro failed after that...I have tried different BLA12 cards on the DSI but same results. No luck.
If the DSI isn't booting you can run VT100 or Hyper Terminal connected to X51,52? Left socket which connects to remote service box. This will then show the results of Bist and show the item which is failing.
That's the mystery, the DSI seems by all means and standards to work fine...the DSI comes on normally and u can enter patient data etc with the DSI keyboard. I suspect the II cabinet but I am tired now, ready to fl y in a technician and just this over with.
If the Dsi is OK it's probably the Figaro board. Does the II controller display the II mag 0 to 2. It,s the small board at the back of the IITV cabinet
Altima will inspire the trust, loyalty and commitment of our customers through positive relationships and collaboration, top quality equipment services, and a clearly superior value proposition. Call Today (469) 644-7063
mab73
Philips Easy diagnost
July 29, 2013 08:52
Philips Easy Diagnost gives a system error when I try to screen. Otherwise the x-ray works perfect, only screening gives error.
If there is a Philips technician with a service doggle for both Generator and DSI cabinets willing to fly to Kenya I am willing to discuss.
to rate and post a comment
chandra sekeran
re: Philips Easy diagnost
July 30, 2013 08:26
Can you brief the issue in detail. Have you seen any error message?
to rate and post a comment
Muhammad Muhammad
re: Philips Easy diagnost
July 30, 2013 08:27
Hello
What problem with Philips Easy Diagnost Systems, Kindly send us ERROR number.
regards,
muhammadarsalan77@gmail.com
to rate and post a comment
mab73
re: re: Philips Easy diagnost
July 30, 2013 09:09
The xray works but the fluoro gives system error on the scopo user interface...I believe it has something to do with the DSI cabinet. It was working then I was setting up optical disk in the DSI cabinet and the fluoro failed after that...I have tried different BLA12 cards on the DSI but same results. No luck.
to rate and post a comment
Gary Barnes
re: re: re: Philips Easy diagnost
July 30, 2013 02:09
If the DSI isn't booting you can run VT100 or Hyper Terminal connected to X51,52? Left socket which connects to remote service box. This will then show the results of Bist and show the item which is failing.
to rate and post a comment
mab73
re: re: re: re: Philips Easy diagnost
July 30, 2013 06:39
That's the mystery, the DSI seems by all means and standards to work fine...the DSI comes on normally and u can enter patient data etc with the DSI keyboard. I suspect the II cabinet but I am tired now, ready to fl y in a technician and just this over with.
to rate and post a comment
Muhammad Muhammad
re: re: re: re: re: Philips Easy diagnost
July 31, 2013 10:14
Hello
Let me know I finish this work, where is the place kindly updating me.
Regards,
to rate and post a comment
mab73
re: re: re: re: re: re: Philips Easy diagnost
July 31, 2013 12:03
We are in Kenya Nairobi...pls let me know how soon you can come. It is urgent.
to rate and post a comment
Gary Barnes
re: re: re: re: re: re: re: Philips Easy diagnost
July 31, 2013 03:48
If the Dsi is OK it's probably the Figaro board. Does the II controller display the II mag 0 to 2. It,s the small board at the back of the IITV cabinet
to rate and post a comment